This Information Applies To: Agilent G7167A/B, G5668A, G7137A, and G4767A Multisamplers
Issue
The bottom sensing feature in Agilent OpenLab Software is used to detect depths of vials or plates. Material of plates or vials must be able to resist the force of the needle; not all vials are suited.
Items Required
Compatible consumables for bottom sensing feature (Figure 1):
Incompatible consumables for bottom sensing feature (Figure 2):
Steps to Follow
The Vial/Well Bottom Sensing feature allows the needle to detect nonuniform well bottoms and adjusts the depth of the needle position above the detected bottom of the vial or well. You can use this feature combined with the Offset to customize the draw position of the needle.
Right click the sampler > select Method, under Advanced > Needle Height Position, check the box Use Vial/Well Bottom Sensing to turn on this feature (see Figure 3).
The default setting is cleared:
Offset
This option is a vertical offset that enables you to position the needle a specific distance (in mm) away from its standard position. The Offset is calculated from the bottom of the vial/well, with a slight amount of clearance to cover tolerances.
Use a positive offset to raise the needle higher in the vial/well to clear precipitates on the bottom of the vial/well, or to select one phase of a biphasic sample.
Figure 3. Method Setup Parameters in Multisampler
1. Sampler icon, 2. Method, 3. Advanced, 4. Needle Height Position, 5. Offset and Use Vial/Well Bottom Sensing, 6. Ok button
Table 1 shows needle tip position when draw position/ offset = 0
|
54 x 2 mL vials (tray) |
No bottom sensing |
5 mm |
With bottom sensing |
2 mm |
Table 1. Needle Tip Position When Draw Position/ Offset = 0 for Multisamplers
Tips: You may want to turn off Vial/Well Bottom Sensing to increase speed of injection, or to avoid the needle touching the bottom of the well in case a sample precipitate could clog it.
|
Learn how to effectively operate your Agilent samplers: Multisamplers Agilent InfinityLab LC Series user manual on Agilent.com |