This Information Applies To: Agilent Variable Wavelength Detector G1314x and G7114A/B modules.
Issue
As UV deuterium lamps ages, the light intensity decreases, the sensitivity of detection may deteriorate. Replace the lamp when the intensity test fails.
The number of hours of use or burn time on your installed UV lamp in your detector module is explained in the article Where to Check Hours on UV Lamp for Agilent UV Detectors?
A lamp change is also needed when the lamp no longer ignites, resulting in a lamp ignition error message on the detector.
Items Required
Steps to Follow
|
Warning: Risk of lesion by touching hot lamp and sharp metal edges.
If the detector has been in use, the lamp may be hot. Wait for the lamp to cool down before handling. Be careful of sharp RFI sheet metal at the rear of the heater fan. |
|
Warning: UV light exposure hazard.
Make sure that the VWD module is switched off at the front power switch, to avoid accidental lamp ignition attempts while working in lamp cover open. |
Figure 1. G1314B/C with lamp cover screw.
Figure 2. G1314D/E/F and G7114A/B module heater fan cover screw.
|
Caution: Do not touch the glass bulb of the new lamp with your fingers as it may reduce the light output.
|
Figure 5. Reinstall lamp
1. Insert new lamp, 2. Tighten two screws.
Tips: The RFID tags used on Agilent UV lamps was changed to a new type in 2015. A firmware update may be required for the detector module to correctly read both new and old-type RFID tags on UV lamps. For more information, see the article Firmware Update Required if New RFID Tag Readers Used.
|
Learn how to effectively maintain/troubleshoot/operate you Agilent Infinity LC system:
Agilent Infinity LC e-learning paths available from Agilent education |