This Information Applies To: Agilent Detectors with non-RFID tagged UV lamps - G1314A/B/C, G1315A/B, or G1365A/B.
Issue
Reset UV lamp hours via Lab Advisor.
Steps to Follow
Figure 1. System Overview of Lab Advisor Interface
1. Connect button, 2. EMF's icon
Figure 2. Example of EMF page in 1100 Series LC
3. Reset button
Note: See Figure 3 for RFID-tagged UV lamps for 1260 Infinity II LC, 1290 Infinity LC & 1290 Infinity II LC.
New lamps with RFID tags will be automatically set to zero upon installation. Therefore, the manual reset feature is not applicable. There will be no button with a message, "Tagged counter cannot be reset." when the cursor hovers over it.
|
Figure 3. Example of an EMF page in RFID-tagged UV lamps
Learn how to effectively operate your Agilent UV Detectors:
Agilent 1260 Infinity LC Systems with OpenLab CDS Agilent 1260 Infinity LC Systems with OpenLab CDS ChemStation Edition eLearning path available from Agilent education |