This Information Applies To: Agilent Analytical Fraction Collector G1364C/F and G5664A/B
Issue
The delay calibration determines the delay time between the detector and fraction collector in the system. The delay is used to compensate the time a compound needs to travel between the point of detection in the detector and the point of collection in the fraction collector (Figure 1). This time difference allows the calculation of the delay volume.
Figure 1. Scheme explaining Delay Volume Calibration
1. Injection of calibration sample (without column), 2. UV detector, 3. Fraction collector, 4. Delay sensor (FDS), 5. Waste
Background
Linking the delay volume to a pump allows the system to automatically change the delay time whenever the flow rate is changed. This is only valid for time and UV-based systems.
The delay volume is calculated as this formula Vdelay = tdelay * F, where:
For an analytical fraction collector, a zero dead volume union is used.
Items Required
Steps to Follow
Figure 2. Run Delay Volume Calibration in Lab Advisor
1. Service & Diagnostics, 2. Fraction collector, 3. Delay volume calibration, 4. Run button
Figure 10. Modify the detector delay volume
1. Right-click fraction collector, 2. Modify, 3. Detector Delay Volumes, 4. Modify the volume.