Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • -ICP-OES Operation
      • ICP-OES Interlocks and Internal Monitoring
      • Water Quality Recommendations for ICP-OES Samples
      • 5000 Series ICP-OES Status Indicators
      • Exhaust Requirements for the 5000 Series ICP-OES
      • Preparing the Instrument Calibration Solution for the Agilent ICP-OES System
      • How to set up communication between Agilent 5000 series ICP-OES Instruments and Agilent ICP Expert software
      • How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES
      • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments
      • SPS4 Alignment and Calibration Wizard
      • How to Upgrade ICP Expert version 7.5
      • ICP-OES Detector and Instrument Calibration
    • +ICP-OES Maintenance
    • +ICP-OES Troubleshooting

How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES

Created by Paul Gunston Paul Gunston over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 10 months ago

This Information Applies To: Agilent 700 Series ICP-OES instruments 


Issue:

It is important that a torch alignment is performed to ensure that the optics are optimized and are focused on the analytical zone of the plasma. This alignment procedure will ensure maximum sensitivity and signal-to-noise.


It is recommended this procedure is carried out regularly, at least once per week, to ensure optimal performance.

The Agilent standard test solution (p/n 6610030100) can be used as it contains 5 mg/L of Mn required for this calibration.

Steps to follow:

It is recommended that you perform a torch alignment scan using the default worksheet provided, as the software will then determine the optimum viewing positions. This worksheet is called Torch Align.

  1. Open the ICP Expert II software and select the Instrument Setup window (if it is not already open), by selecting the Tool Icon on the main toolbar or by pressing CTRL+I on the keyboard.

  2. Turn on the plasma by selecting the Plasma On Icon on the Main toolbar or pressing SHIFT+F4 on the keyboard.

  3.  Aspirate the standard solution.

  4. Select the Torch Align tab.

  5. Select the Graph Type as follows:

    1.    Intensity for high precision results

    2.    SRBR for best detection limit results

  6. Select the Horizontal for the Orientation and then select the Torch Scan button.

  7. Select the Vertical for the Orientation and then select the Torch Scan button.

The signal (Intensity or SRBR) as a function of the torch viewing position will be displayed. The software determines the highest intensity position in the scan, and the corresponding viewing position value is stored and used for all subsequent measurements. (Figure 1)

Figure 1 - Screenshot of the Torch Align worksheet

The torch alignment calibration is now completed.

Tip  Learn how to effectively maintain your Agilent ICP-OES:  
Guidelines for Trouble Shooting and Maintenance of ICP-OES Systems on Agilent.com

  • 700 series
  • Operation
  • ICP-OES Instruments
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help