ICP-OES Articles Checking for a Blocked ICP-OES Nebulizer
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • +ICP-OES Maintenance
    • -ICP-OES Troubleshooting
      • ICP-OES Torch Overheating
      • Software Status Indicators 5000 Series ICP-OES
      • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections
      • Checking for a Blocked ICP-OES Nebulizer
      • Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
      • How to Power Cycle Agilent 5000 ICP-OES Systems
      • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
      • Agilent ICP-OES Performance Test
      • Bypassing the AVS for the 5000 Series ICP-OES
      • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
      • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Checking for a Blocked ICP-OES Nebulizer

Created by Tina Harville Tina Harville over 2 years ago | Last modified by Carlos Vargas Carlos Vargas 9 months ago

This Information Applies To: Agilent 5000 Series ICP-OES


Issue

Blocked nebulizers are the source of many instrument performance issues including loss of sensitivity and poor precision, especially when running organics, high solids, or high salt concentration solutions. The ICP-OES backpressure Nebulizer test verifies nebulizer performance. Run this test if you notice no mist in the spray chamber, have low sensitivity during analysis, or high % RSDs on your replicates. Blockages and partial blockages may not be identified ny visual identification. 


Steps to follow

  1. Turn on the external exhaust and then the ICP-OES.

  2. Open ICP Expert software.

  3. Install the nebulizer, spray chamber, and torch and ensure that the gas supply is connected and turned on. 

     Tip 
    Tips: Perform this test when the nebulizer is new and compare to the results when testing for a blockage.
  4. On the instrument toolbar, click Dashboard > Nebulizer Select drop-down menu >  and select the appropriate nebulizer or Other if nebulizer is not available in the list.

  5. Click Start to run the test.

  6.  Once the test is finished, click Status > Tests to view the results.

  7. A High backpressure result indicates a partial or fully blocked nebulizer or torch injector tube. To correct:

    1. Replace the nebulizer with a brand new one (preferably) or clean one.

    2. Clean the nebulizer.

    3. Clean the torch injector tube.

  8. A Low backpressure result indicates a leak on the nebulizer supply line, the internal capillary of the nebulizer is cracked or broken, or the nebulizer is not installed correctly. To correct:

    1. Check the connections and reinstall the nebulizer.

    2. Check the nebulizer supply lines for cracks.  

 Tip 
Learn how to effectively troubleshoot your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting  e-learning course available from Agilent education
  • Troubleshooting
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help