Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles Maintaining the Axial Pre-Optic Window
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • -ICP-OES Maintenance
      • AVS 6/7 Tubing Connections and Part Numbers
      • AVS 6/7 Maintenance Checklist
      • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule
      • Replacing the AVS 6/7 Rotor
      • Customizing Agilent ICP-OES Maintenance for Sample Matrix and Sample Numbers
      • Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing
      • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices
      • Cleaning the 5000 Series ICP-OES Torch, Organic Matrices
      • Dismantling and Reassembling the ICP-OES Demountable and Semi-demountable Torches
      • Cleaning the ICP-OES Nebulizer
      • Cleaning the Agilent ICP-OES Cone
      • Maintaining the Axial Pre-Optic Window
      • Cleaning the Agilent ICP-OES Spray Chamber
      • Cleaning the Agilent ICP-OES Snout
      • Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES
      • Cleaning the 5100/5110 ICP-OES Air Filter
      • Routine Maintenance 700 Series ICP-OES Systems
      • How to Change the SPS Autosampler Probe
      • Cleaning the 5000 Series Agilent ICP-OES Water Inlet Filter
    • +ICP-OES Troubleshooting

Maintaining the Axial Pre-Optic Window

Created by Tina Harville Tina Harville over 2 years ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES


Issue

The axial pre-optic window needs checking from time to time to ensure it is clean.

If the axial pre-optic window is dirty, it must be cleaned or replaced to ensure adequate sensitivity in the UV region of the emission spectrum. If the axial wavelength calibration fails for wavelengths less than 190 nm, this may indicate that the axial pre-optic window is dirty.


Steps to follow

Follow the steps in this article to clean or replace the axial pre-optic window:

 Caution 
Caution: Always turn off your instrument before attempting to remove the window and do not operate the instrument while the window is removed.

 Warning 
Warning: Hot Surface. The torch compartment and components in it become extremely hot during instrument operation and remain hot for some time after the instrument has been switched off. Allow the torch compartment to cool for at least five minutes before attempting to remove any component. Use heat-resistant gloves.

To remove the axial pre-optic window Assembly:

  1. Remove the torch and snout. See Figure 1.


    Figure 1. Torch Compartment
    1. Axial pre-optic window assembly, 2. Cone, 3. Cone thumb screws, 4. Snout, 5. Radial pre-optic window assembly, 6. RF coil, 7. Torch

  2. Remove the three cone thumb screws.

  3. Pull down the Cone from the assembly.

  4. Gently pull down on the axial pre-optic window assembly.

  5. Tilt the top of the assembly towards the back, and then slide it to the right to remove from the torch compartment. (See Figure 2.)


    Figure 2. Removing the axial pre-optic window assembly
  6. Check the axial pre-optic window, and replace or clean as needed.

To clean the pre-optic window

  1. Soak the optics window in a dilute 5% v/v laboratory detergent for 30 minutes.
  2. Thoroughly rinse with de-ionized water.
  3. If the contamination is not removed, soak the pre-optics window in dilute 5% v/v nitric acid for 1 hour.
  4. Thoroughly rinse with de-ionized water.
  5. Rinse with clean, reagent grade isopropyl alcohol 2-propanol.
  6. Dry the pre-optics window using one of the following options:
    • Air dry at room temperature
    • Dry in a drying oven set between 40–80 °C 104–176 °
    • Use a clean, filtered compressed gas such as argon, nitrogen, or compressed air.
 Caution 
Caution: Do not use hydrofluoric acid to clean the pre-optics window. Always wear gloves when handling the pre-optics window. Avoid touching the faces of the window. Handle the pre-optics window by the edges only. Avoid wiping the faces of the window to prevent scratching the window.

 Warning 
Warning: Chemical hazard: The cleaning procedure may involve the use of corrosive or other hazardous chemicals. Many acids, including nitric acid, are corrosive and can cause severe burns when they come into contact with the body. It is essential that appropriate protective clothing be worn always when handling all chemicals.

To install the axial pre-optic window assembly:

  1. Tilt the top of the axial pre-optic window assembly towards the back of the instrument and then slide it up into the housing. See Figure 3.


    Figure 3. Installing the Axial pre-optic window Assembly

     Caution 
    Caution: Do not hit the RF coil. The axial pre-optic window assembly must be replaced before fitting the cone. The cone will not fit properly if the axial pre-optic window assembly is not installed.
  2. Align the holes in the assembly with the screws and then push the assembly in.

  3. Install the cone onto the assembly.

  4. Tighten the three thumb screws firmly by hand.

 Tip  Learn more on how to effectively troubleshoot your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting e-learning course available from Agilent education
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • Maintenance
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help