Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • +ICP-OES Maintenance
    • -ICP-OES Troubleshooting
      • ICP-OES Torch Overheating
      • Software Status Indicators 5000 Series ICP-OES
      • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections
      • Checking for a Blocked ICP-OES Nebulizer
      • Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
      • How to Power Cycle Agilent 5000 ICP-OES Systems
      • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
      • Agilent ICP-OES Performance Test
      • Bypassing the AVS for the 5000 Series ICP-OES
      • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
      • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS

Created by Whitney Blackmon Whitney Blackmon 3 months ago | Last modified by Carlos Vargas Carlos Vargas 22 days ago

This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7, ICP Expert version 7.X


Issue

The AVS can contribute to instrument performance issues such as sensitivity and precision. During troubleshooting, bypassing (or skipping) the AVS and running a worksheet during can identify if the AVS is the root cause of the issue. Other root causes could be the method application, or an instrument fault. 


Items Required

If an internal standard line is required, install the internal standard tubing on the peristaltic pump and locate a Y-piece connector (p/n 1610132400).

Resolution / Steps to Follow

  1. To bypass the AVS, see Bypassing the AVS for the 5000 Series ICP-OES .
  2. Connect the Y-piece connector to the internal standard tubing, sample tubing, and add a short length of tubing from the Y splitter fitting to the nebulizer (Figure 1):


    Figure 1. Connection of Y-piece for Internal Standard
    1. Y Splitter, 2. Internal standard tubing, 3. Sample tubing, 4. Mixed solution to nebulize

  3. Place the internal standard line in the internal standard solution and the sample line in DI water. Close and adjust the peristaltic pump roller arms.
  4. Turn on the peristaltic pump at normal pump speed (via the Instrument Page in ICP Expert). Confirm the solution in the sample and internal standard lines are being delivered to the nebulizer at a smooth steady rate. Adjust the peristaltic pump roller arm tension if needed.
  5. Check that the spray chamber is draining into the rinse vessel, then turn off the peristaltic pump (via the Instrument Page in ICP Expert).
  6. Create a worksheet from the template used to run the analysis with the AVS. Select ICP Expert > File > New From > select your template > Open (Figure 2):


    Figure 2. ICP Expert home page

  7. Select the Configuration tab > deselect the AVS 6/7 check box (Figure 3):


    Figure 3. Deselect the AVS 6/7

  8. Select the Conditions tab, edit the parameters (Figure 4):
    1. Common Conditions > Uptake delay (s). The uptake delay will depend on the length of the autosampler tubing and the nebulizer capillary. It is recommended to use a stopwatch to time how long it takes for the sample to reach the nebulizer with the peri-pump speed set to ‘Fast pump’.
    2. Common Conditions > Rinse time (s). The rinse time will depend on the length of the autosampler tubing, length of the nebulizer capillary and the identity of present elements and concentrations. If you suspect carry-over, increase the rinse time.
    3. Measurements Conditions > Stabilization time (s). The stabilization time will need to be greater than it was when using the AVS. Set the stabilization time to a minimum of 15 sec. 
       Tip 
      Tip: The recommended stabilization time must be long enough for the solution to reach the plasma. When in doubt, increase the time settings on the Conditions page
       


      Figure 4. Conditions tab

  9. Select the Sequence tab and confirm the standard and/or sample locations (Figure 5):


    Figure 5. Sequence
    tab. Verify that rack tube matches sample/standard tube placement on the autosampler

  10. Save the worksheet and click Run to start the analysis (Figure 6):
     Caution 
    Caution: Save changes made to the worksheet as a new file name to avoid changes to the worksheet set up for AVS use


    Figure 6. Save changes and run the worksheet

  11. If this worksheet completes with acceptable results, an AVS component is the most likely root cause of the poor instrument performance. See the AVS part replacement topics in the ICP Expert Help, How To Section for instructions on replacement (Figure 7):
     Tip 
    Tips: For more AVS maintenance information, see Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule 



    Figure 7. ICP Expert Help

 Tip 
Learn how to effectively operate your 5000 Series ICP-OES:
Agilent 5000 Series ICP-OES: Introduction to ICP Expert Software (Ver. 7.5) e-learning course available from Agilent education

  • Troubleshooting
  • AVS 4
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • ICP OES Accessories
  • AVS 6/7
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help