ICP-OES Articles Bypassing the AVS for the 5000 Series ICP-OES
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • +ICP-OES Maintenance
    • -ICP-OES Troubleshooting
      • ICP-OES Torch Overheating
      • Software Status Indicators 5000 Series ICP-OES
      • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections
      • Checking for a Blocked ICP-OES Nebulizer
      • Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
      • How to Power Cycle Agilent 5000 ICP-OES Systems
      • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
      • Agilent ICP-OES Performance Test
      • Bypassing the AVS for the 5000 Series ICP-OES
      • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
      • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Bypassing the AVS for the 5000 Series ICP-OES

Created by Whitney Blackmon Whitney Blackmon over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 9 months ago

This Information Applies To: 5000 series ICP-OES (5100,5110,5800,5900), AVS 4/6/7


Issue

Steps to bypass the AVS to troubleshoot poor performance on the ICP-OES


 Tip 
Tip: These steps should be completed after the Subsystem Communication, Air Flow, Water Flow, Gas Flows, RF Generator, Camera, and Optics Tests have been completed with a status of ‘Done’ or ‘Pass’

Bypass (or skip) the AVS to run a worksheet and troubleshoot instrument performance issues such as sensitivity and precision. The root cause may be related to an AVS component, the method application, or an instrument fault and bypassing the AVS helps identify the root cause.

With the AVS 6/7 installed on the 5000 series ICP-OES, the sample introduction area will look like Figure 1 (AVS 6) or Figure 2 (AVS 7). The setup difference between the AVS 6 and AVS 7 is the seventh port on the AVS 7 valve for the internal standard. An internal standard tube must be installed on the peristaltic pump.


Figure 1. AV6 connections
1 To Autosampler, 2 To carrier, 3 To waste


Figure 2. AVS 7 connections
1 To AVS, 2 To waste

Steps to follow

  1. Disconnect the UniFit sample connector from the nebulizer. Unscrew the threaded flangeless fitting from port 2 of the AVS, remove, and set this fitting aside. It will not be used when bypassing the AVS (Figure 3):


    Figure 3. UniFit sample connected to the nebulizer and Port 2 of the AVS

  2. Disconnect the barbed end (labeled “Peri Pump Rinse”) of the T-piece from the rinse tubing on the peristaltic pump (Figure 4):


    Figure 4. Peri-Pump rinse connector 

  3. If an internal standard line is installed on the AVS 7, unscrew the threaded internal standard fitting from port 7 of the AVS and remove the tubing from the peristaltic pump. Set the internal standard tubing and connection aside. It will not be used when bypassing the AVS (Figure 5):


    Figure 5. Internal Standard Line connector

  4.  Locate the UniFit sample connector, 0.75 mm id, for U-Series SeaSpray concentric nebulizer. This connector is included with a new nebulizer (Figure 6):

     Tip 
    Tip: This connector is also available to order as a pack of 10 (Agilent part number G8010-80035).


    Figure 6. UniFit connector

  5. Attach the UniFit sample connector to the nebulizer and insert the capillary tube end into the rinse tubing on the peristaltic pump (Figure 7):


    Figure 7. Connector and Rinse Tubing connected

  6. Remove the autosampler threaded fitting from port 5 on the AVS that connects the autosampler probe tubing to the valve (Figure 8):


    Figure 8. Port 5 of the AVS

  7. Locate the ICP-OES connection kit, Autosampler to Peristaltic Pump (p/n G8010-60348). Find the small rectangular block with a barb on one end and a threaded port on the other end. If you:
    1. Have this block: Attach the autosampler threaded fitting into the threaded port of the small rectangular block and insert the barb into the free end of the rinse tubing on the peristaltic pump. This peristaltic pump tubing will no longer act as a rinse line but will now be the sample line and will deliver the solution to the nebulizer, bypassing the AVS valve.
    2. Do not have this block: Pull the threaded autosampler connection fitting off the autosampler capillary and insert the autosampler capillary end into the free end of the rinse tubing on the peristaltic pump. This peristaltic pump tubing will now be the sample line and will deliver the solution straight to the nebulizer, bypassing the AVS valve. It is recommended to have spare autosampler connection fittings available (p/n G8010-60306).
  8. Clamp the peri-pump tubing, send the autosampler to a vial containing DI or rinse, turn on the peristaltic pump, and confirm that the solution is being delivered to the nebulizer at a smooth steady rate. Confirm that the spray chamber is draining. Adjust the roller arm tension for the sample and drain if needed. Turn off the peristaltic pump when completed.

  9. Perform an ICP-OES ICP-OES Detector and Instrument Calibration 
 Tip 
Learn how to effectively troubleshoot your Agilent AVS7 on the 5000 Series ICP-OES:
Agilent 5000 Series ICP-OES Maintenance and Tips for Optimal Performance e-learning course available from Agilent education
  • Troubleshooting
  • AVS 4
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • ICP OES Accessories
  • AVS 6/7
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help