ICP-OES Articles Exhaust Requirements for the 5000 Series ICP-OES
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • -ICP-OES Operation
      • ICP-OES Interlocks and Internal Monitoring
      • Water Quality Recommendations for ICP-OES Samples
      • 5000 Series ICP-OES Status Indicators
      • Exhaust Requirements for the 5000 Series ICP-OES
      • Preparing the Instrument Calibration Solution for the Agilent ICP-OES System
      • How to set up communication between Agilent 5000 series ICP-OES Instruments and Agilent ICP Expert software
      • How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES
      • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments
      • SPS4 Alignment and Calibration Wizard
      • How to Upgrade ICP Expert version 7.5
      • ICP-OES Detector and Instrument Calibration
    • +ICP-OES Maintenance
    • +ICP-OES Troubleshooting

Exhaust Requirements for the 5000 Series ICP-OES

Created by Whitney Blackmon Whitney Blackmon over 1 year ago | Last modified by Whitney Blackmon Whitney Blackmon over 1 year ago

This Information Applies To: Agilent 5000 Series ICP-OES (5100/5110/5800/5900)


Issue

Correct ventilation is required to ensure proper function of the ICP-OES


Resolution 

The ICP-OES must be located under a flue that is vented by an exhaust fan and ducted to an external vent.

 Caution 
Caution: The exhaust system installation must comply with any rules and/or regulations that may be imposed by the local authorities responsible for control of facilities and fixtures in the work place

The exhaust system (flue, ducting, and external vent) requirements are:

  • The ventilation ducting must have an ID of 150 mm (5.9 inches)
  • Minimum flow: 2.5 m3 /min (90 ft3/min) at 2.4 m/s (7.7 ft/s)
  • Maximum flow of 6.0 m3 /min (200 ft3/min) at 5.7 m/s (18.6 ft/s)
  • Exhaust flow must be continuous as long as the plasma is on, and stable with a maximum fluctuation of ±5% of flow
  • The fan control switch and running indicator lamp should be located in a position where the instrument operator can view the indicator and access the control switch
  • Flexible ducting must be used for easy removal during instrument maintenance
  • Ducting must be corrosion resistant, fire-proof and should be kept clear of fire alarms, sprinkler heads, and other heat-sensitive devices
  • The external vent must be fitted with a back draft damper and the outlet location must be clear of doors, windows, and heater or air-conditioning units
  • It is recommended to directly connect the exhaust ducting directly to the 150 mm diameter exhaust vent of the instrument
  • If using an extraction hood, it is important to closely couple the exhaust vent of the instrument to the extraction port with a distance of no more than 1.5 cm (0.6 inches) above the exhaust outlet
 Warning 
Warning: The exhaust system must be on before the plasma is ignited otherwise damage to the instrument or injury to the operator may occur

 Caution 
Caution: Exceeding the exhaust maximum flow could impact the sustainability of the plasma resulting in an error with the instrument function

Please see Figure 1 for 5100/5110 model diagram, and Figure 2 for 5800/5900 model diagram

 

 

Figure 1: 5100/5110 model ICP-OES 

 

Figure 2: 5800/5900 model ICP-OES

 Tip 
Tips: The component parts or a kit required for an exhaust system may be purchased from Agilent: External Inlet Duct Adaptor Kit p/n: G8010-68002

 Caution 
Caution: The customer is responsible for supplying the ductwork between the instrument and the lab extraction system, if not purchased with the instrument.  To allow for personal preferences, control gear switch and pilot light assemblies are not supplied by Agilent.

 Tip 
Learn how to set up the laboratory environment for your ICP-OES:
Agilent 5100/5110 ICP-OES Site Prep Guide or Agilent 5800/5900 ICP-OES Site Prep Guide user manual on Agilent.com

 Tip 
Learn how to effectively operate your ICP-OES:
ICPOES-5900-1200e - Agilent 5800/5900 ICP-OES with ICP Expert: Instrumentation Overview e-learning course available from Agilent education

  • Operation
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help