Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • -ICP-OES Maintenance
      • AVS 6/7 Tubing Connections and Part Numbers
      • AVS 6/7 Maintenance Checklist
      • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule
      • Replacing the AVS 6/7 Rotor
      • Customizing Agilent ICP-OES Maintenance for Sample Matrix and Sample Numbers
      • Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing
      • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices
      • Cleaning the 5000 Series ICP-OES Torch, Organic Matrices
      • Dismantling and Reassembling the ICP-OES Demountable and Semi-demountable Torches
      • Cleaning the ICP-OES Nebulizer
      • Cleaning the Agilent ICP-OES Cone
      • Maintaining the Axial Pre-Optic Window
      • Cleaning the Agilent ICP-OES Spray Chamber
      • Cleaning the Agilent ICP-OES Snout
      • Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES
      • Cleaning the 5100/5110 ICP-OES Air Filter
      • Routine Maintenance 700 Series ICP-OES Systems
      • How to Change the SPS Autosampler Probe
      • Cleaning the 5000 Series Agilent ICP-OES Water Inlet Filter
    • +ICP-OES Troubleshooting

Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule

Created by Jo Boden Jo Boden 9 months ago | Last modified by Jo Boden Jo Boden 5 months ago

This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900)


Issue:

The ICP-OES and AVS 6/7 components require regular routine maintenance. Agilent-authorized field service engineers must perform any maintenance procedures not listed in this document.

 Warning 
Warning: Shock Hazard. Do not remove instrument covers that expose electrical circuits, devices, and components operating at dangerous voltages. Contact with these circuits, devices, and components can result in death, cause serious injury, or painful electric shock.

Maintenance frequency:

 Tip  Recommended maintenance schedule: The maintenance schedule can be customized to suit your sample throughput and sample matrix via the User Maintenance Counters. Customization can reduce system contamination, increase sample throughput and reduce consumable expenses. 

Maintenance tasks are recommended at different intervals:

  • Frequently
  • Daily
  • Weekly
  • Monthly
  • Three monthly

 Run the ICP-OES Performance Test after completing routine maintenance. 

 Tip  Tip: Failure to properly maintain the ICP may lead to sensitivity issues. See Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections.

Frequently

  • Check and empty the drain vessel.

Daily

  • ICP-OES
    • Check the water level in the Argon Humidifier before every use if applicable.
    • Clean the surface of your ICP-OES and accessories. Spills should be cleaned up immediately.
    • Inspect the torch for cleanliness and clean as necessary. Separate processes are used for cleaning acid and aqueous matrices and organic matrices.
    • Inspect the pump tubing and replace if it has lost its elasticity. Unclip the pump tubes when the pump is not in use.
  • AVS 6/7 switching valve
    • Inspect the switching valve tubing if in use. Replace if necessary.
    • Check that the sample loop is filling completely for the first few samples in the first run of the day. If the loop still does not fill completely, the rotor may need to be replaced. 
    • Check that there are no blockages in the tubing, sample probe, or connectors used.
    • Check for blockages in the switching valve. Clean or replace the valve.
    • Check for increases in carry over in blanks. If carry over is increasing, the valve may require the stator to be cleaned or replaced.
    • Flush the sample flow path, including the switching valve assembly, at the end of each period of operation. Flush systems analyzing acid and water samples with de-ionized water, and for systems running organic samples use an appropriate solvent, such as isopropyl alcohol propan-2-ol.

Weekly

  • Clean the torch. Separate processes are used for cleaning when analyzing acid and aqueous matrices, or organic matrices.
  • Clean the cone.
  • Clean the snout.
  • Clean the spray chamber.
  • Clean the nebulizer.

Monthly

  • Clean the cooling air inlet filter on top of your ICP-OES.
  • Perform an ICP-OES instrument calibration.
  • Inspect the state of the RF induction coil. Contact your Agilent representative if maintenance is required.
  • Inspect the external gas supply system for leaks including the tubing connected to the instrument. Replace any damaged, leaking, or worn components.
  • Inspect the axial and radial pre-optics window for cleanliness. Clean or replace as necessary.
  • Check/clean the heat exchanger radiator on the cooling system and remove any build-up of dust and dirt.
  • Check the water level in the water cooler. See the manual supplied with the water cooler for details.
  • Periodically, drain the coolant from the cooling system and then refill/ treat with an appropriate algaecide as recommended by the manufacturer.

Three monthly

  • Remove and clean the water filter on the right side of the instrument.
 Tip 
Learn how to effectively your Agilent ICP-OES and AVS 6/7:
ICPOES-MULTI-2200e - Agilent 5000 Series ICP-OES Maintenance and Tips for Optimal Performance e-learning course available from Agilent education

  • Maintenance
  • ICP OES Accessories
  • AVS 6/7
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help