Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles ICP-OES Interlocks and Internal Monitoring
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • -ICP-OES Operation
      • ICP-OES Interlocks and Internal Monitoring
      • Water Quality Recommendations for ICP-OES Samples
      • 5000 Series ICP-OES Status Indicators
      • Exhaust Requirements for the 5000 Series ICP-OES
      • Preparing the Instrument Calibration Solution for the Agilent ICP-OES System
      • How to set up communication between Agilent 5000 series ICP-OES Instruments and Agilent ICP Expert software
      • How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES
      • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments
      • SPS4 Alignment and Calibration Wizard
      • How to Upgrade ICP Expert version 7.5
      • ICP-OES Detector and Instrument Calibration
    • +ICP-OES Maintenance
    • +ICP-OES Troubleshooting

ICP-OES Interlocks and Internal Monitoring

Created by Paul Gunston Paul Gunston 11 months ago | Last modified by Carlos Vargas Carlos Vargas 8 months ago

This Information Applies To: Agilent 5000 Series ICP-OES Instruments


Issue

The spectrometer incorporates interlocks and covers that are used to monitor critical operations to ensure Operator and instrument protection. If the instrument is used in any manner not specified by Agilent, this protection may be impaired.


5800 ICP-OES

 Warning 
Warning: No interlock or cover must ever be bypassed, damaged or removed. It is good practice to develop safe working habits that do not depend upon the correct operation of the interlocks for safe operation. 

Torch Compartment Door Interlock

  The torch compartment door interlock monitors the state of the internal interlock. An open interlock switch will inhibit plasma ignition or extinguish an existing plasma.

Torch Loader Interlock

The torch loader interlock monitors the state of the internal interlock on the torch loader handle. An open interlock switch will inhibit plasma ignition or extinguish an existing plasma.

Instrument Temperature, Gas Flow and Pressure Monitoring

  The ICP-OES monitors a number of internal temperatures, gas flows and pressures. If these values fall below or exceed specified limits the instrument will inhibit plasma ignition or extinguish an existing plasma.

Water Flow Monitoring

  A water flow switch is fitted on the water manifold assembly. If there is insufficient water flow an error message 'Low water flow' is displayed in the ICP Expert software, and plasma ignition is inhibited. If this error is detected while the plasma is in operation, an immediate shut down will occur.

Cooling Fan Monitoring

  Cooling fan operation is detected by an air flow sensor. If the air-cooling flow is detected to be insufficient, the instrument will inhibit plasma ignition or extinguish an existing plasma.

Plasma On Monitoring

  An optical detector in the ICP-OES is positioned to detect light emitted by the plasma. If the plasma goes out, an alert message is displayed in the ICP Expert software. If this fault condition is detected while the plasma is on, the RF generator and the gas control unit are shut down.

RF Control Watchdog

  The RF control watchdog effectively controls the plasma generating system. If the watchdog system detects that the plasma generation system is not operating correctly, the high-voltage supply is shut down.

High Voltage Power Supply Failure

  If an overload condition is detected, ignition of the plasma will be inhibited or, if the plasma is already running, the RF power is shut down and the plasma is extinguished.

 Tip 
Tip: It is good practice to develop safe working habits that do not depend upon the correct operation of the interlocks for safe operation. 

 Tip 
Learn how to effectively maintain and operate your Agilent ICP-OES:
ICPOES-5900-1200e - Agilent 5800/5900 ICP-OES with ICP Expert: Instrumentation Overview e-learning course available from Agilent education
  • Operation
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help