ICP-OES Articles Replacing the AVS 6/7 Rotor
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • -ICP-OES Maintenance
      • AVS 6/7 Tubing Connections and Part Numbers
      • AVS 6/7 Maintenance Checklist
      • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule
      • Replacing the AVS 6/7 Rotor
      • Customizing Agilent ICP-OES Maintenance for Sample Matrix and Sample Numbers
      • Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing
      • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices
      • Cleaning the 5000 Series ICP-OES Torch, Organic Matrices
      • Dismantling and Reassembling the ICP-OES Demountable and Semi-demountable Torches
      • Cleaning the ICP-OES Nebulizer
      • Cleaning the Agilent ICP-OES Cone
      • Maintaining the Axial Pre-Optic Window
      • Cleaning the Agilent ICP-OES Spray Chamber
      • Cleaning the Agilent ICP-OES Snout
      • Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES
      • Cleaning the 5100/5110 ICP-OES Air Filter
      • Routine Maintenance 700 Series ICP-OES Systems
      • How to Change the SPS Autosampler Probe
      • Cleaning the 5000 Series Agilent ICP-OES Water Inlet Filter
    • +ICP-OES Troubleshooting

Replacing the AVS 6/7 Rotor

Created by Paul Gunston Paul Gunston over 1 year ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent 5000 Series ICP-OES Systems, AVS 6/7


Issue 

Over time the AVS 6/7 rotor can become worn or damaged by chemicals. Agilent recommends inspecting and/or replacing the rotor every 5000 AVS switches or when analytical results or precision fall outside expected limits.


Steps to Follow:

 Tip 
Tip: Refer to the Help section of the Agilent ICP Expert Instrument Software for further detailed instructions.

 Warning  Warning:
Always wear appropriate PPE when handling AVS 6/7 components and tubing to avoid contact with potentially dangerous chemicals. 
  1. It is not necessary to disconnect the tubing from the AVS valve to replace the rotor however, the tubing can be removed and rinsed when there is suspected system contamination. Figure 1 shows the AVS with the tubing removed. 


    Figure 1. 1 AVS7 with the tubing removed

  2. Disassemble the AVS 6/7. Use the 7/16 inch hex key provided with the valve to remove the AVS body. 

  3. Inspect the contact surface of the rotor for signs of wear and replace if there are visible score marks or damage. AVS 6/7 rotor part number G8494-60002, AVS 6 stator and rotor part number G8494-60003. AVS 7 stator and rotor part number G8495-60003

     

  4. Inspect the reverse face of the stator for signs of wear and replace if there are visible score marks or damage.

     

  5. Re-assemble the AVS7 valve.

     Tip  Tip: When reassembling the AVS, it is critical that each screw is tightened evenly in turn to ensure consistent pressure. If the pressure on the three screws is not even or the screws are too tight / too loose, air leaks may occur.

  6. When finished reassembling the AVS, it is recommended to run the Advanced Valve System Test as detailed in the Help section of the Agilent ICP Expert Instrument Software to check performance of the AVS. 

 Tip 
Learn how to effectively maintain your Agilent AVS:
Agilent 5000 Series ICP-OES with ICP Expert e-learning path available from Agilent education
  • Maintenance
  • ICP-OES Accessories
  • AVS 6/7
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help