Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles SPS4 Alignment and Calibration Wizard
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • -ICP-OES Operation
      • ICP-OES Interlocks and Internal Monitoring
      • Water Quality Recommendations for ICP-OES Samples
      • 5000 Series ICP-OES Status Indicators
      • Exhaust Requirements for the 5000 Series ICP-OES
      • Preparing the Instrument Calibration Solution for the Agilent ICP-OES System
      • How to set up communication between Agilent 5000 series ICP-OES Instruments and Agilent ICP Expert software
      • How to perform a Torch Alignment Calibration on the Agilent 700 Series ICP-OES
      • How to Reset the AVS 6/7 on Agilent ICP-OES Instruments
      • SPS4 Alignment and Calibration Wizard
      • How to Upgrade ICP Expert version 7.5
      • ICP-OES Detector and Instrument Calibration
    • +ICP-OES Maintenance
    • +ICP-OES Troubleshooting

SPS4 Alignment and Calibration Wizard

Created by Paul Gunston Paul Gunston over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 6 months ago

This Information Applies To: Agilent Spectroscopy Systems, SPS4 Autosampler


Issue

The sample probe may be displaced from its correct position due to the probe hitting a sample vial or the sides of the wash reservoir. User calibration allows for minor adjustments to center a misaligned probe using a calibration wizard.


The following discussion is a summary of the steps required to run the Calibration Wizard. Refer to the Help section of the Agilent ICP Expert Instrument Software for the detailed procedure.

 Tip  Tip: The SPS 4 autosampler must be connected directly to the PC for the Calibration Wizard to work correctly.

Before starting this procedure, always check the following

  • The probe is straight and mounted tightly to the Z-axis slide.
  • The spill tray and rack location mat are fitted correctly.
  • Sample tubes sit vertically in the sample rack and are not leaning to one side. The fixed wash reservoir is correctly mounted.
  • The rack configuration is correct.
 Tip  Tip: Close the ICP Expert application software before launching the Calibration Wizard.

To access the Calibration Wizard:

  • Click Start > Agilent > SPS 4 Calibration Wizard. Refer Figure 1.

Figure 1. Calibration Wizard program

Figure 2. Calibration Wizard screen

To perform a user calibration

  1. Ensure that the autosampler power is on.
 Tip 
Tip: The sample racks fitted should be empty and not contain any sample vials. 
  1. Start the Calibration Wizard software. Refer Figure 2.
  2. Select the appropriate COM port from the drop-down list, and then click Connect in the Com Port field.
  3. Click Reset to start the calibration process.
  4. The button labeled Rear Left changes to orange and the sample probe moves to point Rear Left above the far-left sample well in the sample rack in that position. To adjust the probe position at point Rear Left do the following:
    1. Adjust the sample probe height to the top of the sample well using the Up/Down buttons.
    2. Adjust the XY position to the middle of the well using the arrow buttons.
    3. When the position of the probe is centered in the sample well, click Capture.
 Tip 
Tip: The indicator next to the Rear Left button changes from orange to green.
  1. The sample probe then moves to point Rear Right above the far-right sample well in the sample rack in that position, and the indicator next to the button labeled Rear Right changes to orange.
  2. Repeat Step 6a, b, and c for the remaining three positions in the following order: Rear Right, Front Left, and then Front Right.
  3. When all four points have been captured, a confirmation message is displayed asking if you want to write the new calibration values.
    • Click Yes to write the calibration values. The autosampler is adjusted using the new calibration value.
    • Click No if you want to retry. The calibration process is restarted.
 Tip  Tip: After each position has been completed, the autosampler will automatically move to the next position.

After the calibration has been successfully completed, check the positioning of the sample probe by clicking each corner position button, and driving the Z-Axis position down. Confirm that the sample probe is positioned in the center of each well.

If the calibration fails or an error message appears, please contact your Agilent Service representative for assistance.

 Tip 
Learn how to effectively maintain, troubleshoot and operate your Agilent SPS 4 Autosampler and Spectroscopy Systems :
Agilent 5000 Series ICP-OES with ICP Expert e-learning path available from Agilent education
  • Operation
  • SPS 4
  • ICP-OES Accessories
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help