ICP-OES Articles Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • -ICP-OES Maintenance
      • AVS 6/7 Tubing Connections and Part Numbers
      • AVS 6/7 Maintenance Checklist
      • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule
      • Replacing the AVS 6/7 Rotor
      • Customizing Agilent ICP-OES Maintenance for Sample Matrix and Sample Numbers
      • Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing
      • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices
      • Cleaning the 5000 Series ICP-OES Torch, Organic Matrices
      • Dismantling and Reassembling the ICP-OES Demountable and Semi-demountable Torches
      • Cleaning the ICP-OES Nebulizer
      • Cleaning the Agilent ICP-OES Cone
      • Maintaining the Axial Pre-Optic Window
      • Cleaning the Agilent ICP-OES Spray Chamber
      • Cleaning the Agilent ICP-OES Snout
      • Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES
      • Cleaning the 5100/5110 ICP-OES Air Filter
      • Routine Maintenance 700 Series ICP-OES Systems
      • How to Change the SPS Autosampler Probe
      • Cleaning the 5000 Series Agilent ICP-OES Water Inlet Filter
    • +ICP-OES Troubleshooting

Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES

Created by Carlos Vargas Carlos Vargas over 1 year ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent 5800/5900 ICP-OES


Issue:

The instrument requires clean, dry, noncorrosive air for cooling purposes. The air enters the instrument through an air supply vent located beside the chimney on the top of the instrument. The vent is fitted with a dust filter to screen out particulate matter from the surrounding environment. This filter should be replaced regularly to prevent reduced air flow.


Remove the filter:

Figure 1 shows how to remove the air filter.

  1. Turn off the instrument and gas supply.
     Caution
    Caution: Always turn off your instrument before attempting to replace a filter to prevent dust and contaminants entering the instrument.
  2. Remove the filter cartridge from the instrument

  3. Lift the two tabs on the cartridge up and then open the cartridge cover.

  4. Pull the bracket out of the cartridge.

  5. Remove and dispose of the used filter according to local regulations.


    Figure 1: Removing the air filet

Install the new air filter:

Figure 2 shows how to install the air filter.

  1. Push the new filter (Agilent part number G8020-68016) securely into the bracket.
  2. Slide the bracket into the cartridge with the filter side facing the large side opening.
  3. Slide the two bracket tabs into the cutouts on the cartridge.
  4. Close the cartridge cover.
  5. Fully insert the cartridge into the instrument with the filter side facing the back of the instrument.



Figure 2: Installing the air filter

Before using the instrument:

  1. Turn on the external exhaust if off.
  2. Turn on the instrument and gas supply.
  3. Purge the instrument for at least 60 minutes.
 Tip 
Recommended maintenance: Agilent recommends replacing the air filter monthly.

 Tip 
Order your new Agilent air filter here

  • ICPOES
  • Maintenance
  • ICP-OES
  • how-to-article
  • ICP Expert
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help