Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles Agilent ICP-OES Performance Test
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • +ICP-OES Maintenance
    • -ICP-OES Troubleshooting
      • ICP-OES Torch Overheating
      • Software Status Indicators 5000 Series ICP-OES
      • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections
      • Checking for a Blocked ICP-OES Nebulizer
      • Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
      • How to Power Cycle Agilent 5000 ICP-OES Systems
      • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
      • Agilent ICP-OES Performance Test
      • Bypassing the AVS for the 5000 Series ICP-OES
      • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
      • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

Agilent ICP-OES Performance Test

Created by Jo Boden Jo Boden over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 17 days ago

This Information Applies To: Agilent ICP-OES 5000 Series (5100/5110/5800/5900)


Issue

How to run the ICP-OES Performance Test after installation, preventative maintenance, repairs, and when troubleshooting.


Steps to Follow

The ICP-OES Performance Test runs a suite of tests including Resolution, Sensitivity, and Precision to determine the analytical performance of your instrument and sample introduction components. The test uses the wavelength calibration solution and a blank to create a one-point calibration, then reruns the wavelength calibration solution as a sample.  

 Tip 
Tips: If 'Advanced Valve System Test' and 'Use Autosampler' are selected on the Tests tab, the Performance Test is not available

Prepare the test solutions:

 Warning
Warning: Wear appropriate Personal Protective Equipment when using acids
  • Blank: 1% nitric acid solution in ASTM grade water Or Calibration blank solution-Ready to use (Agilent part number 5190-7001).
  • Purchased wavelength calibration solution - diluted (Agilent part number ICM-630) Or Purchased ICP-OES wavelength calibration solution-Ready to use (Agilent part number 6610030100).
  • Rinse: 1% nitric acid solution in ASTM grade water.

Prepare the instrument 

Ensure that the instrument meets the following guidelines before performing the Performance Test:

  • Configured with the standard sample introduction system: SeaSpray nebulizer, double-pass cyclonic spray chamber, and 1.8 mm id standard torch. 
  • The torch, nebulizer, spray chamber, and pump tubing are clean and in good working condition. 
  • The external instrument exhaust is operating.
  • The ICP Expert software is open.
  • The polychromator heater and boost gas have been on for at least 20 minutes before lighting plasma.
  • The plasma has been on for at least 20 minutes before running the test. 
  • The Peltier cooling has been on for at least five minutes before running the test. 

Running the Performance Test

  1. Perform a dark current scan.
  2. Perform a wavelength calibration.
  3. Measure the Argon ratio and confirm it is within ±0.5 of the factory setting.
  4. Click Instrument > Tests > and deselect all tests except Instrument Performance
  5. Samples can be analyzed from the autosampler or introduced manually:
    1. Autosampler: select Use Autosampler and then place the wavelength calibration, rinse, and blank solutions in the locations as per the worksheet.
    2. Introduce solutions as prompted.
  6. Click Run Tests. This function automatically fills the appropriate method parameters and begins the test. 
  7. Follow the onscreen prompts to introduce solutions if you are manually introducing the solutions.
  8. To print the report, click Print Preview > Print.

If the instrument performance tests pass, the system is in good working order. If any of the tests fail, contact Agilent Technical Support for additional troubleshooting.

  Tip
Learn how to effectively maintain your Agilent ICP-OES:
ICPOES-EX-1130e - Agilent 5000 Series ICP-OES: Introduction to ICP Expert Software (Ver. 7.5) e-learning course available from Agilent education




  • Operation
  • Troubleshooting
  • 5900
  • ICP-OES Instruments
  • 5100
  • 5110
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help