Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles How to Power Cycle Agilent 5000 ICP-OES Systems
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • +ICP-OES Maintenance
    • -ICP-OES Troubleshooting
      • ICP-OES Torch Overheating
      • Software Status Indicators 5000 Series ICP-OES
      • Loss of or Change in Agilent ICP-OES Sensitivity: Causes and Corrections
      • Checking for a Blocked ICP-OES Nebulizer
      • Cleaning a Blocked Nebulizer in the Agilent ICP-OES System
      • How to Power Cycle Agilent 5000 ICP-OES Systems
      • Troubleshooting Failed Plasma Ignition for Agilent 5000 Series ICP-OES
      • Agilent ICP-OES Performance Test
      • Bypassing the AVS for the 5000 Series ICP-OES
      • Set up an ICP-OES Worksheet with Internal Standard while Bypassing the AVS
      • Troubleshooting ICP-OES and ICP-MS Chillers (G3292A and G8481A)

How to Power Cycle Agilent 5000 ICP-OES Systems

Created by Whitney Blackmon Whitney Blackmon over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 1 month ago

This Information Applies To: The Agilent 5000 ICP-OES Series and Agilent ICP Expert Software


Issue

How to reset the Agilent ICP-OES system when experiencing a communication error between the instrument and software, or if the instrument seems to be experiencing issues when trying to light the plasma or run a worksheet. These steps will help to determine if there is a hardware issue with the instrument.


Steps to Follow 

 Caution 
Caution: Save all data in ICP Expert software before shutting down.

Before performing the following steps, save and close any data or methods that are open and turn off the plasma (if on).

  1. Turn off the water chiller. Turn off the chiller by pressing the power button on the front of the unit, then switch off the circuit breaker switch on the back of the unit.

  2. Close the software and shut down the PC:

    1. Select File >Exit after saving your data in the Agilent ICP Expert Software, or select the Exit Icon on the ICP Expert launch screen (Figure 1)

    2. Select the Windows option to Shutdown the PC (right click the Windows icon). 

      Figure 1: ICP Expert Software Exit Icon

  3. Turn off any autosamplers connected to the ICP-OES. If using an Agilent SPS 4 Autosampler, turn off the SPS 4 by pressing the power switch located near the power cord.

  4. Turn off the ICP-OES. Press and release the lower left front Standby button (Figure 2). Wait until the instrument status LED (upper right side) has a steady flash, or is off. This should take a minute or two, depending on model of ICP-OES. Then press the rocker switch (Power) located on lower left rear side of the instrument next to the power cord (Figure 3) to position "0" (Off).

     Figure 2- Standby button on front panel

    Figure 3 - Power rocker switch

  5. Allow the instrument to power completely down and remain in an off state for five to ten minutes before proceeding.

  6. When powering on the system, the order will be reversed.

  7. Turn on the ICP-OES:

    1. Press the lower left rear side switch (Power) on the ICP-OES to position "1". Wait a minute or two for the instrument to start the powering on process.

    2. Press the lower left front button (Standby) on the ICP-OES.

    3. Allow the instrument to fully initialize before proceeding. Once the instrument status LED (upper right side) is a flashing amber color, proceed to the next step.

  8. Turn on the autosampler:

    1. Press the power button on the SPS 4 autosampler, and allow it to go through initialization.

    2. The probe should move to all four corners of the autosampler and return to the rinse station.

  9. Turn on the PC and Software:

    1. Turn on the PC

    2. Open the ICP Expert software.

    3. If the software does not automatically connect to the instrument, select the Instrument page option, then select Connect and choose the ICP-OES from the drop-down menu. If more than one instrument is set up for your software, please choose the corresponding ICP-OES that has been restarted.

    4. Successful instrument connection can be verified with a solid green upper right side instrument status LED on the ICP-OES.

The ICP-OES system has now been restarted, and if no errors are present, proceed as normal. If errors are present, please contact Agilent Technical Support.

 Tip 
Learn how to effectively operate your Agilent ICP-OES:
ICPOES-5110-1200e - Agilent 5110 ICP-OES Instrumentation Overview
ICPOES-5900-1200e - Agilent 5800/5900 ICP-OES with ICP Expert: Instrumentation Overview
e-learning courses available from Agilent education
  • Troubleshooting
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help