Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles Cleaning the Agilent ICP-OES Spray Chamber
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • -ICP-OES Maintenance
      • AVS 6/7 Tubing Connections and Part Numbers
      • AVS 6/7 Maintenance Checklist
      • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule
      • Replacing the AVS 6/7 Rotor
      • Customizing Agilent ICP-OES Maintenance for Sample Matrix and Sample Numbers
      • Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing
      • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices
      • Cleaning the 5000 Series ICP-OES Torch, Organic Matrices
      • Dismantling and Reassembling the ICP-OES Demountable and Semi-demountable Torches
      • Cleaning the ICP-OES Nebulizer
      • Cleaning the Agilent ICP-OES Cone
      • Maintaining the Axial Pre-Optic Window
      • Cleaning the Agilent ICP-OES Spray Chamber
      • Cleaning the Agilent ICP-OES Snout
      • Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES
      • Cleaning the 5100/5110 ICP-OES Air Filter
      • Routine Maintenance 700 Series ICP-OES Systems
      • How to Change the SPS Autosampler Probe
      • Cleaning the 5000 Series Agilent ICP-OES Water Inlet Filter
    • +ICP-OES Troubleshooting

Cleaning the Agilent ICP-OES Spray Chamber

Created by Agilent Knowledge Portal Agilent Knowledge Portal over 1 year ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent 5100, 5110, 5800, 5900 ICP-OES systems


Issue

The ICP-OES spray chamber filters the sample droplets created by the nebulizer. Larger droplets fall to the chamber drain, while the smaller droplets are introduced to the gas flow going into the ICP torch plasma. A dirty spray chamber causes large droplets to form, which leads to incomplete transfer of sample to the torch, causing poor precision and memory effects. 


ICP-OES overview 

The ICP-OES sample introduction components are detailed in Figure 1.

   
Figure 1: ICP-OES sample introduction components
1. Torch loader handle 2. Torch 3.Torch clamp 4. Nebulizer gas connections 5. Spray chamber drain tubing 6. Spray chamber 7. Nebulizer 8. Nebulizer inlet tubing

How to identify a dirty spray chamber

 Tip Routine analysis of aqueous and nonaqueous samples using a single spray chamber will increase cleaning and instrument downtime. Agilent recommends the use of two spray chambers, one for aqueous solutions and one for organic solutions, to reduce downtime. 

A clean spray chamber has a uniform mist on the internal surface (Figure 2). A dirty spray chamber has large droplets on the internal surfaces (Figure 3).


Figure 2: Correctly wetting spray chamber

 
Figure 3: Incorrectly wetting spray chamber

Steps to follow

  1. Turn off the pump and plasma.
     Warning
    Warning: The spray chamber may contain potentially dangerous chemicals, and the nitric acid and alkali detergent wash solutions are corrosive. Always wear Personal Protective Equipment when cleaning the spray chamber.
  2. Remove the nebulizer (see Figure 1).
  3. Remove the spray chamber and disconnect the tubing. This video demonstrates how to remove the spray chamber:
  4. Rinse the spray chamber with 1% HNO3 solution then ASTM Type 1 water. 
  5. Prepare an aqueous solution of a quality laboratory alkaline detergent. Completely immerse the spray chamber in the solution, ensuring all the internal surfaces of the spray chamber are in contact with the detergent solution.
  6. For:
    • Single pass spray chambers: The container may be placed in an ultrasonic bath and sonicated for 40 minutes.
    • Multiple pass spray chambers: Do not sonicate, instead allow the spray chamber to stand overnight in the detergent solution.
       Caution 
      Caution: Sonicating a multiple pass spray chamber may cause damage.
  7. Remove the spray chamber from the detergent solution and rinse thoroughly with ASTM Type 1 water.
  8. Dry external surfaces using a clean, lint free paper towel. Do not dry the spray chamber using compressed gases.
  9. Re-attach all tubes, adaptors, and the nebulizer. Reconnect the assembly to the instrument.
  10. Test the cleaned spray chamber by nebulizing a blank solution and repeat the cleaning process if there are large droplets on the spray chamber internal surfaces.
     Tip 
    Tips: It is not unusual that some spray chambers will require extended cleaning.
  11. If the spray chamber is not going to be used immediately after cleaning:
    • Store in an appropriate covered vessel containing a 1% HNO3 solution until required or,
    • Store the dried spray chamber in its original packaging.
Tip Recommended maintenance schedule:  Agilent recommends weekly to monthly cleaning of the spray chamber. Analysis of samples with complex matrices may require more frequent cleaning. 

 Tip 

Learn how to effectively maintain your Agilent ICP-OES:
ICPOES-5110-1270e - Agilent 5110 ICP-OES Maintenance and Troubleshooting  e-learning course available from Agilent education

  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • Maintenance
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help