Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
ICP-OES Portal
  • Agilent Knowledge Portal
ICP-OES Portal
ICP-OES Articles Cleaning the ICP-OES Nebulizer
  • ICP-OES Articles
  • ICP-OES Videos
  • Atomic Spectroscopy Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -ICP-OES
    • +ICP-OES Operation
    • -ICP-OES Maintenance
      • AVS 6/7 Tubing Connections and Part Numbers
      • AVS 6/7 Maintenance Checklist
      • Recommended Routine ICP-OES and AVS 6/7 Maintenance Schedule
      • Replacing the AVS 6/7 Rotor
      • Customizing Agilent ICP-OES Maintenance for Sample Matrix and Sample Numbers
      • Replacing and Adjusting the 5000 Series ICP-OES Peristaltic Pump Tubing
      • Cleaning the 5000 Series ICP-OES Torch, Acid and Aqueous Matrices
      • Cleaning the 5000 Series ICP-OES Torch, Organic Matrices
      • Dismantling and Reassembling the ICP-OES Demountable and Semi-demountable Torches
      • Cleaning the ICP-OES Nebulizer
      • Cleaning the Agilent ICP-OES Cone
      • Maintaining the Axial Pre-Optic Window
      • Cleaning the Agilent ICP-OES Spray Chamber
      • Cleaning the Agilent ICP-OES Snout
      • Cleaning the Air Filter of the Agilent 5800/5900 ICP-OES
      • Cleaning the 5100/5110 ICP-OES Air Filter
      • Routine Maintenance 700 Series ICP-OES Systems
      • How to Change the SPS Autosampler Probe
      • Cleaning the 5000 Series Agilent ICP-OES Water Inlet Filter
    • +ICP-OES Troubleshooting

Cleaning the ICP-OES Nebulizer

Created by Agilent Knowledge Portal Agilent Knowledge Portal over 1 year ago | Last modified by Carlos Vargas Carlos Vargas over 1 year ago

This Information Applies To: Agilent 500 Series ICP-OES (5100, 5110, 5800, 5900)


Issue

The ICP-OES nebulizer converts the liquid sample into an aerosol as it enters the spray chamber. A dirty or partially blocked nebulizer traps sample, which leads to poor reproducibility and reduced sensitivity. Droplets on the internal surface of the spray chamber can indicate poor conversion of liquid to mist.


ICP-OES overview

The ICP-OES sample introduction components are detailed in Figure 1. 

 
Figure 1: ICP-OES sample introduction components
1. Torch loader handle 2. Torch 3.Torch clamp 4. Nebulizer gas connections 5. Spray chamber drain tubing 6. Spray chamber 7. Nebulizer 8. Nebulizer inlet

Tip Tips: Not all nebulizers are the same, and fittinmgs will vary with the nebulizer.

Blocked nebulizer

A blocked nebulizer can be identified by a visual inspection as per Figure 2. 


Figure 2: Blocked nebulizer
1. Blockage at tip 2. Sample trapped in the nebulizer tip

Steps to follow

Caution Caution: The concentric glass nebulizer is fragile. Handle with care. Do not place a glass nebulizer in an ultrasonic bath as vibrations may cause edge chipping or cracking and degrade the performance of the nebulizer. Do not use hydrofluoric acid with glass or quartz sample introduction components.
  1. Turn off the pump and the plasma. Remove the nebulizer, being careful not to touch the tip of the nebulizer.

     
  2. Different cleaning solutions can be used to clean the nebulizer including 2.5% detergent solution, methanol, or 10% nitric acid. Flush the nebulizer with the selected cleaning solution in both directions:

    1.  Push liquid through using a syringe attached with a small piece of soft plastic tubing or the nebulizer cleaning tool (part number G3266-80020). This tool is suitable for concentric nebulizers. It cannot be used with the Agilent OneNeb nebulizer as it differs in construction. 



    2. Pull liquid through with a syringe attached with a small piece of soft plastic tubing to the nebulizer inlet. Block the nebulizer gas inlet side arm. Place the tip of the nebulizer in the liquid and draw the liquid through the sample inlet end. 

    3. Rinse with de-ionized water.

  3. If the blockage remains, use one or more of the following cleaning steps:
    1. Attach a gas hose to the side arm of the nebulizer. Gently flush with laboratory grade air or nitrogen.
    2. Soak the nebulizer overnight in concentrated nitric acid. Use a Pasteur pipetted to gently fill the central capillary then submerse the entire nebulizer. Ensure that no air is trapped in the capillary.
    3.  To remove salts, soak the nebulizer in 25% detergent overnight. Use a Pasteur pipetted to gently fill the central capillary then submerse the entire nebulizer. Ensure that no air is trapped in the capillary.Warning
      Warning Warning: Wear Personal Protective Equipment when handling nitric acid, methanol, and laboratory detergents.
  4. Rinse well with water and allow to dry before use.
  5. Reinstall the nebulizer. 

Tip Recommended maintenance schedule:  Agilent recommends weekly cleaning of the nebulizer. Analysis of samples with complex matrices may require more frequent cleaning. 

 Tip
Learn how to effectively maintain your Agilent ICP-OES: 
ICPOES-5110-1270e, Agilent 5110 ICP-OES Maintenance, and Troubleshooting e-learning course available from Agilent education


  • Troubleshooting
  • 5900
  • 5800
  • ICP-OES Instruments
  • 5100
  • 5110
  • Maintenance
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help