GC/MS Articles How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • +GC/MS Operation
    • -GC/MS Maintenance
      • How to Open the Side Plate of the GC/MSD
      • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)
      • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
      • How to Disassemble and Assemble the CI Source
      • How to Clean the EI/CI Source in the Agilent 5977A/B
      • How to Disassemble and Assemble the HES (High Efficiency Source)
      • How to Clean the HES (High Efficiency Source)
      • How to Replace the Electron Multiplier Horn - 5977 Series GC/MSD
      • How to Bake Out an Agilent GC/MSD
      • Foreline Pump Oil change for GC/MSD Systems
      • How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
      • Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
      • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
    • +GC/MS Troubleshooting

How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD

Created by Xu Li-hua Xu Li-hua over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 8 days ago

This Information Applies To: Agilent GC/MS MassHunter Acquisition Software Agilent 5975/5977 GC/MSD


Issue

Use the Early Maintenance Feedback (EMF) tool to track the GC/MSD's consumable resource and part use, and to configure alerts for when they need to be replaced or serviced. After certain maintenance activities, for example, replacing the foreline pump oil, it is necessary to reset the EMF counters.

There are two methods to set or reset the EMF counters.


Steps to follow

Method 1

Using the Macro function, see the following steps:

  1.  In the GC/MS Acquisition software command line, type: macro “emfcheck.mac”, go and press [Enter] in the keyboard (see Figure 1). 


    Figure 1. Agilent GC/MS Acquisition software command line

  2. The Select EMF action dialog box is displayed (see Figure 2). Select one of the displayed option buttons according to the desired task: Set Limits, Set Counters or Reset Counters, and click OK. Select any of the checkboxes to execute the previously selected action, and click OK.
    In this example, to reset the Pump oil (Days) to zero, select Reset Counters > Pump oil (days) 0 of 90 (see Figure 3). 


    Figure 2. Select EMF action dialog box 


    Figure 3. Reset EMF counters dialog box

  3. To disable an EMF function, set the EMF parameter limit value to -1. In the following example, to disable the EMF Pump Oil (Days) function, select Set Limits from the Select EMF action dialog box (Figure 2). Click OK, then type -1 in the Pump Oil (days) 0 of 90 box (see Figure 4). 


    Figure 4. Enter limits for EMF parameters dialog box

Method 2

If the instrument is an Agilent 5977B GC/MSD with LVDS communication enabled, there is another option to set the EMF, see the following steps:

  1. In the MassHunter Acquisition software, from the Instrument menu, select Maintenance (see Figure 5)

     
    Figure 5. Agilent GC/MS MassHunter Instrument menu

  2. The Early Maintenance dialogue box is displayed, select Early Maintenance Feedback Counters > Mass Spec > Pump maintenance (see Figure 6).

     
    Figure 6. Early Maintenance dialog

  3. To expand the pump maintenance information, click the Plus sign (+) on the right side.

  4. Click Reset EMF to reset the pump oil days count to zero in this example. The Warning Threshold and Service due Threshold can also be enabled and configured in this dialog box (They are disabled as default) (see Figure 7).

     
    Figure 7. Agilent MassHunter Acquisition software Early Maintenance Feedback Counters

 Tip 
Learn how to effectively operate your Agilent MassHunter software:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning path available from Agilent education
  • Troubleshooting
  • MassHunter Acquisition Software
  • GCMS Instruments
  • GCMS Data Acquisition
  • single quadrupole
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help