Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
GC/MS Articles Using a Column Installation Tool to Install a Capillary Column on Agilent 5977A/B GC/MSD
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • -GC/MS Operation
      • How To Pump Down (Start) the GC/MSD - EI - MassHunter
      • How To Vent (Shut Down) the GC/MSD - EI - MassHunter
      • How to Install a Capillary Column on 5977 Series GC/MSD
      • Using a Column Installation Tool to Install a Capillary Column on Agilent 5977A/B GC/MSD
    • +GC/MS Maintenance
    • +GC/MS Troubleshooting

Using a Column Installation Tool to Install a Capillary Column on Agilent 5977A/B GC/MSD

Created by Carlos Vargas Carlos Vargas over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh over 1 year ago

This Information Applies To: Agilent GC/MS 5977A, 5977B 


Content

This article shows how to install a capillary column on 5977A/B GC/MS using a column installation tool.


Steps to follow: 

Warning Warning: – The GC and MS operate at high temperatures. To avoid burns, do not touch any parts of GC and MS until you are sure they are cool.
– Always wear safety glasses when handling capillary columns. Use care to avoid puncturing your skin with the end of the column.
– Always wear clean gloves while handling any parts that go inside the GC or the analyzer chambers.

1. To vent the MS, loosen the vent valve by turning the knob counterclockwise gently until you hear air being drawn into the chamber (see Figure 1). For more information, please see How To Vent (Shut Down) the GC/MSD - EI - MassHunter.

 
Figure 1. GC/MS vent valve

2. Use a standard column nut to install a capillary column into the GC inlet (Capillary column installation quick reference guide). Alternatively, you could use a Self Tightening column nut (part number G3440-81011).

3. Condition the column.

4. Slide a standard column nut and conditioned ferrule onto the free end of the GC column. The tapered end of the ferrule must point towards the nut (see Figure 2). Slide the column into the column installation tool (see Figure 3, part number G1099-20030). Make sure the column end projects a specified distance past the end of the column installation tool. It is best to push the column the whole way through, cut the column, then pull back until it’s at a specified length. See the following information for the specified lengths for each type of source.

  • For an SST/Inert/Xtr EI source installation, the column extends 1 to 2 mm past the end of the column installation tool. 

  • For a High Efficiency Source (HES) installation, the column extends 4 to 5 mm. 

  • For a CI Source Installation, the column extends about 1 mm.

 
Figure 2.  1-Ferrule  2-Standard column nut

 
Figure 3. GC/MSD column installation tool

5. Alternatively, you could also use a Self Tightening column nut (see Figure 4, part number G3440-81013) with a short ferrule to install a column. See Figure 5 for ferrule part numbers.

 
Figure 4.  Self-tightening column nut for MS

Column ID  Ferrule p/n
0.1–0.25 mm 5181-3323
0.32 mm 5062-3514
0.45–0.53 mm 5062-3512

Figure 5. Short graphite/vespel (15%/85%) ferrules

6. Take out the column with the nut from the column installation tool, slide it into the GC/MS interface, hand-tighten the nut. Ensure that the position of the column does not change as you tighten the nut, and be sure to not over tighten the nut. if you use a self tightening nut, tighten the nut in the clockwise direction, continue to tighten until you feel the ferrule grip the column.

7. Check the GC oven to be sure that the column does not touch the oven walls. Column installing steps are finished here for a self tightening nut. Continue steps 8 and 9 for a standard column nut.

8. Tighten the nut 1/4 to 1/2 turn.

9. Check the nut’s tightness after one or two heat cycles and retighten as required.

Tip

Learn more on how to effectively operate your GC/MS:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning path available from Agilent education. 

  • single quadrupole
  • ei
  • 5977B
  • ci
  • 5977A
  • Maintenance
  • GC/MS Instruments
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help