GC/MS Articles How to Open the Side Plate of the GC/MSD
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • +GC/MS Operation
    • -GC/MS Maintenance
      • How to Open the Side Plate of the GC/MSD
      • How to Clean the EI/CI Source in the Agilent 5977A/B
      • How to Clean the HES (High Efficiency Source)
      • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)
      • How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
      • Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
      • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the CI Source
      • How to Disassemble and Assemble the HES (High Efficiency Source)
      • How to Bake Out an Agilent GC/MSD
    • +GC/MS Troubleshooting

How to Open the Side Plate of the GC/MSD

Created by Carlos Vargas Carlos Vargas over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh 8 months ago

This Information Applies To: Agilent 5975 and 5977 Series GC/MSD


Issue:

For some GC/MSD maintenance or repair asks, it is required to open the side plate (filament exchange, ion source cleaning, electron multiplier exchange, column installation).

This article describes how to open the side plate, or what to do when the side plate cannot be opened.


Steps to follow: 

1. Vent the system, see: How To Vent (Shut Down) the GC/MSD - EI - MassHunter 

2. Remove the analyzer covers (see Figure 1).


Figure 1. Opening the MSD covers.

3. Disconnect the side board control cable and source power cable from the side panel. See the location of the cables in Figure 2.

4. Loosen the analyzer side plate thumbscrews if they are fastened. See the location of the cables in Figure 2.


Figure 2. Side plate cables and screws
1. Thumbscrews, 2. Source Power Cable, 3.Side Board Control

5. To open the analyzer, gently swing the side plate out. See Figure 3 for a short video on steps 3 to 5.


Figure 3. Opening the Analyzer

When the instrument is vented and turned off, the side plate is not easy to pull apart. Remember to open the vent valve to equilibrate the pressure. After some time to allow the analyzer chamber to equilibrate, the side plate should be easily opened. If you still cannot open it in a reasonable time, you can carefully pry it from the lower right side of the side panel with a tool such as a flat screwdriver from the location shown in Figure 4.

 Caution 
Caution: Be careful not to damage or gouge the side plate and sealing ring when inserting tool. Only use this method as a last resort after trying all other suggestions mentioned in this article.


Figure 4. Open from location circled in red, as last resort.

 Tip 
Learn how to effectively operate and troubleshoot your Agilent 5977 Series GC/MSD:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning path available from Agilent education.
GCMS-5977-1270r - Agilent 5977B GC/MSD Diagnostics and Troubleshooting e-learning course available from Agilent education. 
  • single quadrupole
  • Best practices
  • 5977B
  • 5977A
  • Maintenance
  • GC/MS Instruments
  • 5975
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel