Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
GC/MS Articles Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • +GC/MS Operation
    • -GC/MS Maintenance
      • How to Open the Side Plate of the GC/MSD
      • How to Clean the EI/CI Source in the Agilent 5977A/B
      • How to Clean the HES (High Efficiency Source)
      • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)
      • How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
      • Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
      • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the CI Source
      • How to Disassemble and Assemble the HES (High Efficiency Source)
      • How to Bake Out an Agilent GC/MSD
    • +GC/MS Troubleshooting

Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD

Created by Carlos Vargas Carlos Vargas over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh 8 months ago

This Information Applies To: Agilent 5975 and 5977 series GC/MSD


 Issue: 

If the PFDTD signal is absent or low, but the reagent ions are normal, you may have run out of CI Calibrant. 

Inspect the calibration vial on the back of the flow controller. If necessary, fill the vial with PFDTD according to the instructions in this article.


Steps to Follow:

Materials needed

  • Wrench, open-end, 1/4-inch × 5/16-inch (p/n 8710–0510)
  • PFDTD calibrant, for GC/MS – 1 mL (p/n 8500–8510)
  • Syringe or pipette

 CI calibrant refilling process

  1. Set the reagent gad flow to Gas Off.

  2. Vent the MSD.

  3. Turn off the GC.

  4. Remove the capillary column from the GC/MSD interface.

  5. Place a column nut with a blank ferrule on the end of the interface. This action will help keep contamination out of the MSD.

  6. The foreline pump may be located either on the floor, on the lab bench next to the MSD, or under the analyzer chamber at the back of the MSD. Move it as needed to provide slack in the tubing and cables.

  7. Move the MSD away from the GC until you have access to the GC/MSD interface cable. (See Figure 1)

     Caution 
    Caution: Disconnecting the interface, or LVDS cable with the GC on, can cause a fault condition.

  8. Disconnect the GC/MSD interface cable and the LVDS cable (if applicable). LVDS cable position will vary on different models of GC (See Figure 1)


    Figure 1. Separating the MSD and GC
    1.Interface Cable, 2. LVDS cable (if present)

  9. Continue to move the MSD until you have access to the CI Calibration Valve.

  10. Loosen the calibration vial collar by turning it counterclockwise. Do not remove the collar.

  11. Remove the calibration vial. (See Figure 2)


    Figure 2. Removing the CI Calibration Vial

     Caution 
    Caution: Do not rinse the vial with any solvents. Never expose the inside of the vial to chlorinated solvents, isopropyl alcohol, or water. Exposition to the mentioned solvents will result in severe loss of CI sensitivity.

  12. Fill the vial no higher than the bottom of the internal tube with fresh PFDTD calibrant (8500–8510).

  13. Replace the vial and tighten the collar. (See Figure 3)


    Figure 3. Filling and installing the CI Calibration Vial

  14. Reconnect the Interface cable, and the LVDS cable.

  15. Reposition the MSD next to the GC.

  16. Reinstall the capillary column.

  17. Pump down the MSD.

 Caution 
Caution: After removing the calibrant vial, you must purge the calibration valve. Failure to do so will result in severe contamination of the ion source and damage to the filament and electron multiplier.

To Purge the Calibration Valve

After adding new PFDTD to the CI calibration vial, you must purge the air out of the vial and valve.

  1. If the vacuum gauge controller is on, turn it off.
  2. Verify that PCICH4.U file is loaded.
  3. In the Tune and Vacuum Control view, under the Vacuum menu, click Purge Calibrant Valve. This action will open the CI calibration valve for several minutes with all analyzer voltages turned off
 Tip 
Learn how to effectively maintain your Agilent 5977B GC/MSD:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning path available from Agilent education.
  • Operation
  • 5977B
  • ci
  • 5977A
  • Maintenance
  • GC/MS Instruments
  • 5975
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help