GC/MS Articles Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • +GC/MS Operation
    • -GC/MS Maintenance
      • How to Open the Side Plate of the GC/MSD
      • How to Clean the EI/CI Source in the Agilent 5977A/B
      • How to Clean the HES (High Efficiency Source)
      • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)
      • How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
      • Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
      • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the CI Source
      • How to Disassemble and Assemble the HES (High Efficiency Source)
      • How to Bake Out an Agilent GC/MSD
    • +GC/MS Troubleshooting

Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD

Created by jinhoshin jinhoshin over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh 8 months ago

This Information Applies To: Agilent 5975 and 5977 series GC/MSD


Issue:

Tuning the Agilent GC/MSD with EI sources, requires the use of perfluorotributylamine (PFTBA). The calibration vial will be depleted eventually due to regular use. This article explains how to refill the calibration vial.


Steps to follow:

Materials required

  • GC/MS Tuning Standard - PFTBA (p/n 05971-60571)
  • Syringe or pipette

To Refill the Calibrant vial

  1. Stop any tuning or data acquisition.

  2. Turn off the analyzer and the vacuum gauge controller (if installed). (See Figure 1).

    1. In the Tune and Vacuum Control view, under the Execute menu, click MS OFF.

    2. Under the Vacuum menu, click Tune Vacuum Gauge on/off...

    3. Select the off radio button, and click OK.


      Figure 1.
      Turning off vacuum gauge and analyzer

  3. After removing the front analyzer window cover, loosen the calibration vial collar

  4. Without removing the collar, remove the calibration vial. (See Figure 2)



    Figure 2. Removing the EI Calibration Vial

  5. Using a syringe or pipette, fill about 70 µL of PFTBA into the vial (Figure 3). With the vial vertical, the liquid should be just below the end of the internal tube. In the horizontal position, the liquid should not touch the internal tube (Figure 4).


    Figure 3. Filling the Calibration vial


    Figure 4. PFTBA level in the calibration vial.

  6. Push the calibration vial into the vial, as far as possible, and then withdraw the vial 1 mm.

  7. Turn the collar clockwise to tighten it. The collar should be snug but not overly tight.

  8. Reinstall the analyzer window cover (Figure 5).


    Figure 5. Reinstalling the EI Calibration Vial

 Caution 
Caution: After removing a calibration vial, you must purge the calibration valve. Failure to do so may result in damage to the filaments and the electron multiplier.

To purge the calibration valve

  1. If the vacuum gauge controller is on, turn it off.

  2. In Tune and Vacuum Control view, under the Vacuum menu, select Purge Calibrant Valve (Figure 6).


    Figure 6.
    Purging the Calibrant Valve

The EI Calibration will open for several minutes with all analyzer voltages turned off.

 Tip 
Learn how to effectively maintain your Agilent 5977B GC/MSD:
Agilent 5977 GC/MSD with MassHunter Workstation e-learning path available from Agilent education.
  • Operation
  • ei
  • 5977B
  • 5977A
  • Maintenance
  • GC/MS Instruments
  • 5975
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help