Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
GC/MS Articles How to Disassemble and Assemble the CI Source
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • +GC/MS Operation
    • -GC/MS Maintenance
      • How to Open the Side Plate of the GC/MSD
      • How to Clean the EI/CI Source in the Agilent 5977A/B
      • How to Clean the HES (High Efficiency Source)
      • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)
      • How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
      • Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
      • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the CI Source
      • How to Disassemble and Assemble the HES (High Efficiency Source)
      • How to Bake Out an Agilent GC/MSD
    • +GC/MS Troubleshooting

How to Disassemble and Assemble the CI Source

Created by Agilent Knowledge Portal Agilent Knowledge Portal 6 months ago | Last modified by Agilent Knowledge Portal Agilent Knowledge Portal 3 months ago

This Information Applies To: Agilent CI Source


Issue:

During normal operation of the GC/MS, the CI ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or routine replacement of consumable parts (filament and ceramic insulators). The ion source will then need to be assembled prior to re-installing.  Correct abrasive cleaning, assembly, and installation are essential to robust and reliable operation.


Tools Needed:

  • Gloves, clean, lint-free (8650-0030)
  • Cloths, Lint Free (05980-60051)
  • Hex Driver 1.5mm (8710-1570)
  • Tweezers (8710-0007) 

     Figure 1. Parts Breakdown of the CI Source

    Item Number Item Description Part Number
    1 Setscrew G1999-20022
    2 Filament screw G1999-20021
    3 CI interface tip seal G3870-20542
    4 CI repeller insulator G1999-20433
    5 CI lens insulator G3170-20540
    6 CI drawout cylinder G1999-20444
    7 CI drawout plate G1999-20446
    8 CI ion source heater block assembly G3870-60415*
    9 Entrance lens G3170-20126
    10 CI ion source body G3170-20430
    11 CI ion focus lens G1999-20443
    12 CI repeller G1999-20432
    13 CI filament G7005-60072
    14 Dummy filament G1999-60454
    15/16 Curved Spring washer 3050-1374

    * If replacement is required then contact Agilent for repair options



To Disassemble the  CI source:

  1. Remove the ion source and place it on a clean cloth
  2. Use the 1.5mm Hex Driver to remove the gold plated screws and washers securing the CI filament and dummy filament to the CI ion source and remove them. ( See Figure 2 for steps 2-4)
  3. Separate the CI repeller heater-sensor block assembly from the CI ion source body.
  4. Remove the CI repeller and CI ceramic insulator from the heater-sensor block.


    Figure 2
    . Remove the CI filament, dummy filament, CI repeller and CI insulator from the CI heater-sensor block.

  5. Use the 1.5mm Hex Driver to remove the gold plated set screw from the side of the ion source body. ( See Figure 3 for steps 5 - 8)
  6. Pull the entrance lens and ion focus lens to remove them from the ion source body.
  7. Remove the CI drawout cylinder and the CI drawout plate from the CI ion source body
  8. Separate the entrance lens and ion focus lens from the lens insulator.


    Figure 3.
     Remove and disassemble the lens, drawout plate, drawout cylinder and insulators from the source body and heater source block. 

  9. Proceed with (+) How to Clean the EI/CI Source in the Agilent 5977A/B - GC/MS Articles - GC/MS Portal - Agilent Community or assembly if broken insulators or filaments were being replaced.

To Assemble the CI Source:

  1. Insert the CI drawout plate and the CI drawout cylinder into the CI ion source body. (See Figure 4 for steps 1 - 4)
  2. Assemble the ion focus lens, entrance lens, into the lens insulator.
  3. Slide these assembled parts into the CI ion source body.
  4. Install the setscrew that holds the lens insulator in place.


    Figure 4. Assemble the drawout plate, drawout cylinder, lens and insulators into the source body. 

  5. Attach the ceramic disk to the repeller and place on the source body (See Figure 5 for Steps 5 - 7)
  6. Place the heater block assembly on top of the source body.
  7. Install the filament and the dummy filament using the two gold plated screws and spring washers
     Caution 
    Caution: Do not overtighten the repeller nut or the ceramic repeller insulators will break when the source heats up. The nut should only be finger-tight


    Figure  5
    . Assemble the repeller, filament, dummy filament, repeller and insulators into the source body and heater source block. 

 Tip 
Learn how to effectively operate your 5977:
Agilent 5977 GC/MSD with MassHunter Workstation or Agilent 5977 GC/MSD with OpenLab CDS e-learning course on Agilent University
  • GC/MS
  • 5977B
  • SQ
  • 5977A
  • 5975
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help