Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
GC/MS Portal
  • Agilent Knowledge Portal
GC/MS Portal
GC/MS Articles How to Disassemble and Assemble the HES (High Efficiency Source)
  • GC/MS Articles
  • GC/MS Videos
  • GC/MS Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -GC/MS
    • +GC/MS Operation
    • -GC/MS Maintenance
      • How to Open the Side Plate of the GC/MSD
      • How to Clean the EI/CI Source in the Agilent 5977A/B
      • How to Clean the HES (High Efficiency Source)
      • How to Remove and Reinstall 5977 GC/MSD Ion Sources (EI/CI/HES)
      • How to Reset the Pump Oil EMF Counter of the Agilent GC/MSD
      • Refilling the EI Calibration Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the 5977 EI SST and Inert Ion Sources
      • Refilling the CI Calibrant Vial for the Agilent 5975 and 5977 Series GC/MSD
      • How to Disassemble and Assemble the CI Source
      • How to Disassemble and Assemble the HES (High Efficiency Source)
      • How to Bake Out an Agilent GC/MSD
    • +GC/MS Troubleshooting

How to Disassemble and Assemble the HES (High Efficiency Source)

Created by Agilent Knowledge Portal Agilent Knowledge Portal 9 months ago | Last modified by Agilent Knowledge Portal Agilent Knowledge Portal 3 months ago

This Information Applies To: Agilent HES source, Agilent 5977 Series GC/MSD


Issue:

During normal operation of the GC/MSD, the ion source will need to be disassembled to proceed with abrasive ion source cleaning and / or replacement of consumable parts (dual filament assembly and ceramic insulators). The ion source will then need to be assembled prior to re-installing. Proper abrasive cleaning, assembly, and installation are essential to robust and reliable operation.


Tools Needed:

  • Gloves, clean, lint-free
  • Cloths, clean 
  • T6 Torx driver
  • Tweezers
     

    Figure 1.  Pars of the HES Source

  • Item Number Item Description Part Number
    1 Source finger grip G7002-20008
    2 Filament block G7002-20019
    3 Extractor lens (5)*, with 3 mm opening G7004-20061
    4 Ceramic insulator for extractor G7002-20064
    5 Entrance lens assembly, Extended, HES (1)* G7004-20065
    6 Ion focus lens (2)* G7004-20068
    7 Lens insulator/holder G7002-20074
    8 M2 x 0.4 screw x 12 mm long gold plated screw G7002-20083
    9 Source body G7002-20084
    10 Post extractor lens 2 (3)* G7004-20090
    11 Post extractor lens 1 (4)* G7004-20004
    12 M2 x 6 mm gold plated screw G7002-20109
    13 Locking ring lens insulator G7002-20126
    14 High efficiency dual filament G7002-60001
    15 Ring heater/sensor assembly G7002-60058
    16 Source mount 1.5 mm G7002-60053
    17 Repeller assembly G7002-67057

    * The number in parenthesis is the number engraved on the lens



To Disassemble the HES source:

  1. Remove the ion source and place it on a clean cloth
  2. Use a T6 Torx screwdriver to remove the two screws securing the source mount to the source body. (See Figure 2 for steps 2-8)
  3. Use the finger grip to remove the source mount from the lens body.
  4. Use a T6 Torx screwdriver to remove the screw securing the filament block to the source mount and using the finger grip, remove the filament block.
  5. Remove the dual filament from the filament block.
  6. Remove the finger grip from the filament block.
     Caution 
    Caution: Use care when removing the filament from the filament block. Putting excessive stress on the filament can break or crack it. If this happens, do not attempt to operate with a defective filament, it must be replaced.
  7. Remove the repeller and ring heater/sensor assembly from the source body.
  8. Separate the repeller from the ring heater assembly.
     Caution 
    Caution: Use care when removing and handling the ring heater/sensor assembly. Putting excessive stress or pressure on this assembly can break or crack it. If this happens, do not attempt to operate with a damaged ring heater/sensor assembly, it must be replaced.


    Figure 2.  Disassembling the source to remove the source mount, filament block, filament, repeller and ring heater sensor assembly from the source body.

  9. Use a T6 Torx screwdriver to remove the screw and locking ring for the lens insulator that secures the lens stack in the source body, then remove the lens stack. (see Figure 3 for steps 9-10)
    1. If necessary, use gravity to remove the lens stack ceramic insulator from the source body.
  10. Remove the five lenses from the lens insulator/holder.

     Caution 
    Caution: Use care when removing the lenses from the lens insulator casing. Putting excessive stress on this casing can break or crack it. If this happens, do not attempt to operate with a defective lens insulator, it must be replaced.



    Figure 3
    . Removing the lens and lens stack from the source body. 

  11. For source cleaning instructions see How to Clean the HES (High Efficiency Source)

To Assemble the HES Source:

  1. Assemble the five lenses inside the lens insulator. The lens number is engraved into the outer circumference of each lens. (See Figure 4 for steps 1-3)
     Caution 
    Caution: Use care when inserting the lenses into the lens insulator casing. Putting excessive stress on this casing can break or crack it. If this happens, do not attempt to operate with a defective lens insulator, it must be replaced.
  2. Starting with the entrance lens 1, set the lens into the end groove in the lens insulator and rotate the lens until you feel the ball seat into the circular recess. 
  3. Insert the next four lenses, in numerical order, into the lens insulator.


    Figure 4
    . Assembly of lens into the lens insulator. The lens chamber's open end always faces the entrance lens 1. Rotate each lens until you feel the ball seat into the circular recess. It's easier to insert lens 5 on an angle because the lens stack at this point makes the insulator less flexible. 

  4. Insert the extractor ceramic insulator into the source body (see Figure 6 for steps 4-6)
  5. Insert the lens stack encased in the insulator into the source body. The engraved numbers face the full opening extending to the end of the source body. Verify that the ceramic is sitting flush at the end of the source body. (See Figure 5)


    Figure 5. Ensure that the grooves are aligned when inserting the lens. 

  6. Using the T6 Torx driver, install and secure the gold plated setscrew and lock ring lens insulator that holds the lens assembly in place



    Figure 6
    . Assembly of lens insulator into the source body.

  7. Place the heater/sensor assembly over the guide pins on the source body with the four electrical pins oriented down the flat side of the source body. (See Figure 7 for steps 7- 13)
  8. Place the repeller onto the heater/sensor assembly with the flat side of the repeller circumference aligned with the interface socket in the source body.
  9. Place the source mount over the repeller.
  10. Finger tighten the two gold plated screws using a T6 Torx screwdriver to secure the source mount to the source body.
     Caution 
    Caution: Do not overtighten the screws in the source body. Damage may occur to the repeller.
  11. On the side of the filament block opposite the finger grip, orient the dual filament ceramic holder so that it lines up with the flat of the filament block. Fully insert the three filament leads into the filament block.
  12. Inset the filament block into the source holder and use a T6 Torx screwdriver to secure it to the holder with the gold plated screw.


    Figure 7.
     Assembling the filament, filament block, heater/sensor assembly, and repeller into the source body. 

 Tip 
Learn how to effectively operate your 5977 or 7000 or 7010:
Agilent 5977 GC/MSD with MassHunter Workstation
Agilent 7000 or 7010 Triple Quadrupole GC/MS with MassHunter Workstation 
e-learning paths available from Agilent education

  • GC/MS
  • single quadrupole
  • 5977B
  • 5977A
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help