GC Articles Checking for External Leaks on 7697A Headspace
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • +Basic Troubleshooting for GC Systems
      • Using the Condensation Test to Troubleshoot the GC Inlet
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
      • Gathering GC Diagnostics Log Files Using the Agilent Acquisition Software
      • Gathering Diagnostics Log Files from the 8890/8860/Intuvo 9000 GC Systems Using the Browser Interface
    • +GC Best Practices

Checking for External Leaks on 7697A Headspace

Created by Agilent Knowledge Portal Agilent Knowledge Portal over 2 years ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: Agilent 7697A Headspace Sampler


Issue:

There is a potential leak external to the Headspace unit that can cause loss of pressure.


Steps to Follow:

This article is a troubleshooting guide for when you suspect there is an external leak in the headspace sampler (HS) when it is installed on the inlet of a GC. (Figure 1)

Figure 1: External Leaks can occur in multiple locations: Gas supply bulkhead fittings, Gas cylinder fittings, Regulator fittings, Traps, Shut- off valves (if installed, not included in ship kit), Unions and tee fittings (if installed, not included in ship kit), transfer line connection to the GC

 Warning 
Warning: A pressure decay test cuts off carrier flow through the HS and GC. To prevent damage to the flow path components (column, inlet liner, and so forth), cool the HS and GC before beginning.
  1. Cool the HS sample loop, transfer line, and GC inlet and column oven to prevent damage during the test.
  2. Turn off the HS.
  3. Set the regulator pressure to 415 kPa (60 psi).
  4. Fully turn the regulator pressure adjustment knob counterclockwise to shut the valve.
  5. Wait 5 min.
    1. If there is a measurable drop in pressure, there is a leak in the connections or an internal leak. Utilize a leak detector to locate any external leaks. 
    2. No drop in pressure indicates that the external connections are not leaking 
 Tip 
Learn how to effectively maintain your 7697A Headspace Sampler:
SI-7697-2100sV3 - Agilent 7697A Headspace Sampler Operation and Maintenance e-learning available from Agilent education.

  • Troubleshooting
  • Headspace Sampler
  • 7697A
  • GC Sample Introduction
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help