Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Capillary Column Installation Quick Reference Guide
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • +GC Troubleshooting
    • -GC Best Practices
      • Capillary Column Installation Quick Reference Guide

Capillary Column Installation Quick Reference Guide

Created by Xu Li-hua Xu Li-hua over 1 year ago | Last modified by kristen.sensabaugh kristen.sensabaugh 8 months ago

This Information Applies To: All Agilent Gas Chromatographs, 6890, 6850, 7890, 7820, 8890, and 8860.


Issue:

A quick reference guide on the capillary column installation for the GC.

Inlets Detectors

Split/splitless inlet
Purged packed inlet
Multimode Inlet
Cool on Column (COC) Inlet
PTV Inlet
Volatiles Interface (VI) Inlet

FID / NPD
TCD - Standard Low Leakage Ferrules
TCD - Original Design Column Adapter
µECD
FPD


Steps to follow

 Warning 
Warning: The oven and/or inlet/detector may be hot enough to cause burns. Cool the GC down before column installation or if either is hot, wear heat-resistant gloves to protect your hands.

 Tip 
Tips: Wear clean, lint-free gloves to prevent contamination of parts with dirt and skin oils.

Inlets

Split/splitless inlet

  • Place a septum over the column, then the column nut and ferrule. Trim the end of the column with a column cutter.

  • Pull the column back so that 4-6 mm of the column is extending past the end of the ferrule (see Figure 1).

  • Thread the column nut and column into the inlet and tighten slightly past where the column grabs – retighten after heating. 


Figure 1
1. Ferrule, 2. Column nut, 3. Septum 


Purged packed inlet

  • Place a septum over the column, then the column nut and ferrule. Trim the end of the column with a column cutter.

  • Pull the column back so that 1-2 mm of the column is extending past the end of the ferrule (see Figure 2).

  • Thread the column nut and column into the inlet and tighten slightly past where the column grabs – retighten after heating.


Figure 2

Multimode Inlet

 Tip 
Tips: Make sure that the column adapter nut on the inlet base is fully threaded on and spinning freely – Collar Up! (see Figure 3)

Figure 3
  • Position the column so it extends 10-12 mm (for high boilers you can use 12-14 mm) above the end of the ferrule (see Figure 4). 

Figure 4
  • Tighten with two wrenches - 1/4” and 5/16” to prevent damage the inlet threads (see Figure 5).

Figure 5

Cool On Column (COC) Inlet

  • Insert the column all the way into the inlet until you feel the spring tension – do not withdraw.

  • Tighten with two wrenches - 1/4” and 5/16” to avoid damaging the inlet (see Figure 5).

PTV Inlet

  • There should be 17 mm of the column above the Graphpack ferrule – the Graphpack ferrule should be installed with the graphite end towards the inlet base (see Figure 6).

  • The column nut is slotted. Use a 5 mm wrench to tighten the fitting.


Figure 6  

 

Volatiles Interface (VI) Inlet

  • There is a longer column nut for the VI so that you don’t have to remove the inlet block. The column nut part number is p/n G3504-20504.

Figure 7

 

Detectors

FID/NPD 

  • Place a septum over the column, then the column nut and ferrule. Trim the end of the column with a column cutter.

  • Thread the column nut and column into the detector base. Gently insert the column into the detector until it bottoms, then withdraw 1 to 2 mm before tightening the nut.

     Tip 
    Tips: The dimensions shown of 48 and 68 mm are provided for narrow columns (<100 µm) which may slide all the way through the jet tip. (Figure 8)

Figure 8

 

TCD – Using the standard Low Leakage Ferrules

  • Place a 1/8” Swagelok nut and the back and front ferrules over the column. (Be sure to select the front ferrule size to match the column!) Trim the end of the column with a column cutter.

  • Thread the column nut and column into the detector base. Insert the column all the way into the detector jet until it stops, then withdraw 1 to 2 mm before tightening the nut. Retighten after heating.

Figure 9
1. Front ferrule, 2. Back ferrule, 3. Nut

TCD – Original Design Column Adapter

  • Install the column adapter to the 1/8-inch Swagelok TCD base using a graphite/vespel ferrule.

  • Thread the column nut and column into the adapter. Insert the column into the detector until it bottoms, then withdraw 1 to 2 mm before tightening the nut.



Figure 10
1. Ferrule, 1/8-inch, 2. Nut, 1/8-inch, 3. Capillary column adapter

 

µECD

  • Thread the column nut with column into the make-up gas adapter. Insert the column all the way into the mixing liner until it stops, then withdraw 1 to 2 mm - Use 2 wrenches to tighten the nut. Retighten after heating.

  • Be sure that the make-up gas adapter is installed with the 19 mm dimension shown (see Figure 11).


    Figure 11



    Figure 12  
     Tip 
    Tips: The 70 mm dimension shown is provided for narrow columns (<100 µm) which may go all the way through the mixing liner (see Figure 12).

FPD

  1. Install the column into the Column Measuring Tool provided.

  2. Trim the column such that 1 mm extends from the end of the tool.

  3. Tighten the column nut and mark the position of the column at the back of the nut.

  4. Install into the detector base.


Figure 13
1. Score column here, 2. Column measuring tool, 3. Ferrule, 4. Column nut, 5. The column length *

 

 Tip 
Tips: * The column length depends on the GC and FPD type. FPD+ for 7890B/8890/8860 is 137.5 mm, FPD for 7890A, and some 6890 is 145 mm, FPD for some early 6890 is 153 mm.

 

 Tip 
Learn how to effectively install GC capillary columns:
Installation Care and Maintenance of Capillary GC Columns e-learning available from Agilent education.
  • Detectors
  • GC Systems
  • Best practices
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help