GC Articles How to Calibrate the Agilent 7693A Sample Tray
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • -GC Maintenance
      • +GC Systems
      • -GC Samplers
        • Removing and Installing a Syringe on the Agilent 7693A Autosampler
        • How to Align the Agilent 7693A GC Injector
        • Changing the Turret of the 7693A Automatic Liquid Sampler
        • Replacing the Syringe Carriage Assembly in the Agilent 7693A Automatic Liquid Sampler
        • How to Align the Agilent 7693A ALS Sample Tray
        • How to Calibrate the Agilent 7693A Sample Tray
        • How to Align the Agilent 7683B GC Injector
        • Replacing a Broken Transfer Line on 7697A Headspace Sampler
    • +GC Troubleshooting
    • +GC Best Practices

How to Calibrate the Agilent 7693A Sample Tray

Created by Carlos Vargas Carlos Vargas over 2 years ago | Last modified by kristen.sensabaugh kristen.sensabaugh over 1 year ago

This Information Applies To: Agilent 7693A Automatic Liquid Sampler


Issue:

The Automatic Liquid Sampler (ALS) system calibration process aligns the sample tray with the injector turret position to enable vial transfers without incident.

Calibration should be performed if a calibration does not exist, as a routine maintenance procedure and is recommended if any ALS components are moved.


Steps to follow

  1.  Place the calibration vial part number G4514-40588 (Figure 1) in the tray position 1 (Figure 2).

    Figure 1. Calibration Vial G4514-40588

    Figure 2. Tray position 1

  2. Remove any vials from transfer turret positions L1, L2, and L3 for all installed injectors.

  3. Start the ALS system calibration:

    • Intuvo 9000 GC: In the browser interface, or touch screen, select Settings > Calibration > ALS/Tray > Start Calibration.

    • 8890 GC: In the browser interface, or touch screen, select Settings > Calibration > ALS/Tray > Start Calibration.

    • 8860 GC: In the browser interface, select Settings > Calibration > ALS/Tray > Start Calibration.

    • 7890 Series GC: Use the GC front keypad and select [Options] Calibration > ALS > Start Calibration.

    • 6890N Series GC: Use the GC front keypad and select [Options] Calibration > Sample tray > Start Calibration.

    • 6890A/Plus Series GC: Use the GC front keypad and select [CONFIG] > Sample tray > Calibration > [START]

  4. The following calibration process will occur for all installed injectors:

    1. The tray assesses the turret alignment by placing the calibration vial into turret position L1 and returning to sample tray position 1.

    2. The tray tests the vial height and turret position using the alignment tab between positions L1 and L2 on the transfer turret.

    3. The tray verifies the turret alignment by placing the calibration vial into turret position L1 and returning it to sample tray position 1.

  5. When the calibration process is complete, the green Ready status light will come on and the gantry will stop in the Home position. Remember to replace any vials that were moved for the calibration process.

 

 Tip 
Learn how to effectively operate your Agilent Automatic Liquid Sampler:
SI-7693-1100s - Agilent 7693A ALS Basic and Advanced Operation and Maintenance e-learning available from Agilent education.
  • Automated Liquid Sampler
  • 7693A
  • Maintenance
  • GC Sample Introduction
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help