Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
    • +GC Best Practices

How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems

Created by Vincent Nguyen Vincent Nguyen 6 months ago | Last modified by Carlos Vargas Carlos Vargas 22 days ago

This Information Applies To: Agilent GC Systems with Split/Splitless Inlet (S/SL)


Issue

The Split/Splitless Inlet (S/SL) can accumulate contaminants when high boiling point and/or high molecular weight compounds are present in the injected sample. Often these compounds are present from the sample matrix. The accumulated  contamination is not effectively removed using the analysis method parameters. These accumulated contaminants may appear in the chromatogram as ghost or artifact peaks. They can also be retained on the inside surface of the inlet liner, gold seal, or inlet weldment increasingly affecting the performance of the inlet over time.


Background

During the S/SL inlet bake-out procedure (Steps 1 to 7), high boiling point and/or high molecular weight compounds are subjected to higher inlet temperature and flow in an attempt to vaporize and remove these contaminants. The contaminant molecules, if vaporized, are purged out of the inlet under high split flow to the split vent. A bake-out of column and detector are in the subsequent steps (Steps 8 to 12) of the procedure.

Steps to Follow

 Warning 
Warning: The oven, inlet, and detector may be hot enough to cause burns. Wear heat-resistant gloves to protect your hands.
  1. Create and save a new method for the bake-out procedure. 
  2. Set the inlet into split mode. 
  3. Set the column flow to the normal operating value, or set the capillary column gas velocity to ~30 cm/s.
  4. Set the inlet split vent flow to 200 mL/min by adjusting the split ratio. For example, if the column flow is 2 ml/min, set the split ratio to 100:1. 
  5. Set the inlet temperature to 300 °C, or 25 °C above the normal operating temperature (do not exceed 350 °C as it may thermally damage the inlet liner O-ring).
  6. Make sure to send to the instrument the settings from steps 2 to 5 before continuing.
  7. Purge the column with carrier flow for at least 15 minutes before heating the oven.
  8. Open Method Editor in the software and proceed to the next step. 
  9. If the column is attached to the detector, set the detector 25 °C above normal operating temperature (do not exceed 350 °C). If the column is not attached to the detector, cap the detector fitting.

     Caution 
    Caution: Heating the oven/column up before adequate purging time with carrier gas could damage the column since residual oxygen in the carrier gas, when heated will damage the stationary phase of the column. 
    Do not set the inlet, oven/column, or detector temperature above the column being used manufacturers maximum isothermal operating temperature.
  10. Set the column oven 25 °C above the GC method final oven temperature to bake contaminants from the column. Do not exceed the column manufacturer's maximum temperature limit.
  11. Bake out for 30 minutes or until the detector baseline is stable, and free of contamination peaks.
  12. Return the instrument to operating condition by downloading the analytical method from the software to the instrument. 

If this S/SL bake-out procedure does not restore the inlet performance, replacement of the inlet liner and gold seal will likely be required as the accumulated contamination has not been totally removed. This would be typical of samples with matrix compounds that are nonvolatile but soluble in the injected sample solvent.

 Tip 
Learn how to effectively troubleshoot your GC Split/Splitless Inlet:
GC-7890-2245s - Agilent 7890 Series GC: Split/Splitless Inlet Troubleshooting e-learning course available from Agilent education

  • Troubleshooting
  • S/SL
  • 7820
  • Intuvo 9000
  • GC Systems
  • Maintenance
  • 7890
  • 8890
  • 8860
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help