Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Configuring the Flame Ionization Detector Makeup Gas for Agilent 7890 GC Systems
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • -GC Operation
      • How to Use the Browser Interface for 8890 and 8860 GC Systems
      • Cleaning Local Data Storage From 8890 or 8860 GC Systems
      • How to Lock/Unlock GC Columns with Agilent OpenLab CDS
      • Agilent GC Detector Sampling (Data) Rate
      • Configuring the Flame Ionization Detector Makeup Gas for Agilent 7890 GC Systems
    • +GC Maintenance
    • +GC Troubleshooting
    • +GC Best Practices

Configuring the Flame Ionization Detector Makeup Gas for Agilent 7890 GC Systems

Created by Carlos Vargas Carlos Vargas 2 months ago

This Information Applies To: Agilent GC Systems with Flame Ionization Detector (FID)


Issue

This article will describe how an operator can configure the makeup gas on both the instrument keypad and the software driver (Method Editor).


Background:

The Flame Ionization Detector (FID) on an Agilent Gas Chromatograph (GC) use three types of gasses to maintain a stable flame and signal output. The three gasses are: Hydrogen as fuel, Air as oxidizer, and either Helium or Nitrogen as the makeup gas. Selection between Helium or Nitrogen is a user choice. 

Resolution

 Warning 
Warning: Hydrogen gas is flammable and explosive.

In the Agilent FID, make-up gas flow from the EPC module and joins the FID hydrogen flow (see Figure 1). The combined gases flow up through the FID jet where they mix with the GC carrier gas flow. This gas mixture then exits the FID jet and then is combined with the FID Air supporting the FID flame at the base of the FID collector.

FID EPC Agilent.png
Figure 1: 7890 GC FID pneumatics diagram. This illustration shows a 7890 GC system, but this configuration is similar to other models of the Agilent GC systems with EPC control of the FID gases. 
1. Hydrogen, 2. Makeup gas, 3. Air, 4. Column, 5. Vent, 6. Frit, 7. Valve, 8. Pressure sensor, 9. Restrictor

Using the instrument keypad

Verify or configure the detector makeup gas type using the 7890A GC Keypad by performing the following steps:

  1.  Press [Config][device], where [device] is one of the following 
    1. [Front Det]
    2. [Back Det]
    3. [Aux Detector 1]
    4. [Aux Detector 2]
  2. Scroll to Makeup gas type (or Makeup/reference gas type) and press [Mode/Type]. 
  3. Scroll to the correct gas and press [Enter].

Note: The selected makeup gas type must match the gas supply.

Using the software driver

From the acquisition software Method Editor, the user can select Helium or Nitrogen as a makeup gas (see Figure 2).

Note: The selected makeup gas type must match the gas supply.

FID Makup Gas Configuration in GC Driver - Agilent.png
Figure 2: Method Editor of a 7890A GC on OpenLab CDS ChemStation C.01.10. Please note that the same GC drivers are used on other Agilent GC acquisition software products. 

 

 Tip 
Learn how to effectively operate your Agilent GC FID:
GC-0GEN-1013s - GC Detectors Theory and Operation 
GC-MULTI-2250s - FID Maintenance and Troubleshooting e-learning courses available from Agilent education
  • Operation
  • 6890
  • 7820
  • Detectors
  • 9000
  • 7890
  • 8890
  • 8860
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help