Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • -GC Maintenance
      • -GC Systems
        • Replacing the Inlet Septum for Split/Spitless and Multimode Inlets
        • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC
        • Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System
        • Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System
        • How to Clean an FID
        • Bake Out the Agilent Flame Ionization Detector
        • Bake Out the Agilent Thermal Conductivity Detector
        • Bake Out the Agilent Micro Electron Capture Detector
        • Agilent GC Firmware Update Procedure
        • How To Find More Information About the Agilent Gas Chromatograph micro-ECD Wipe Test Kit
        • How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 GC Systems
        • How To Clear the Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatograph
        • Column Configuration for the 8860/8890 GC System
        • Column Configuration for the 7890 GC System
      • +GC Samplers
    • +GC Troubleshooting
    • +GC Best Practices

Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System

Created by Carlos Vargas Carlos Vargas 5 months ago

This Information Applies To:  Agilent 7890/7820 GC Systems with SSL, MMI, PTV, or VI inlets. 


Issue

Split vent trap replacement is recommended as part of the scheduled preventive maintenance. Sometimes, it has to be replaced when the inlet pressure is too high, or the split vent flow cannot be reached. 

The replacement frequency will vary with the GC application. 

For other GC models, see: Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System 


Items Required

  • QuickPick PM kit, for split vent trap (p/n 5188-6495), includes filter cartridge and two O-rings
  • Torx T-20 key 

Steps to follow

  1. Set the oven temperature and the inlet temperature to 35oC, and allow them to cool down

     Warning 
    Warning: The oven or inlet may be hot enough to cause burns. If either is hot, wear heat-resistant gloves to protect your hands.
     
  2. When the temperatures reach near 35oC, set the total flow to zero and turn off the carrier gas supply pressure
  3. Remove the pneumatics cover to access the trap assembly (Figure 1):

     Warning 
    Warning: The split vent trap may contain residual amounts of any samples or other chemicals you have injected into the GC. Follow your company’s safety procedures for handling these types of substances while replacing the trap filter cartridge.


    Figure 1 Trap assembly.
    1. Split vent valve, 2. Knurled nut on weldment assembly, 3. Retaining clip, 4. Split vent trap screw

  4. Completely loosen the split vent valve screws (Figure 1, 4) that secure the split vent valve in place
  5. Loosen the screw that holds down the split vent trap and rotate the retaining clip
  6. Loosed the knurled nut (counterclockwise) on the weldment assembly. This assembly contains the split vent trap
  7. Lift the split vent valve and trap assembly together from the mounting bracket, and unscrew the split weldment on the filter trap assembly 

     Caution 
    Caution: Be careful not to stress (twist or bend) the tubing between the split vent valve and the trap as it make cause small fractures and leaks. 
        
  8. Open the weldment, remove the trap and the two O-ring seals
  9. Seat the new O-rings on the new filter cartridge
  10. Insert the new trap into the weldment. The narrow end goes toward the front of GC
  11. Start closing the weldment by threading the knurled nut onto the other side. Do not tighten yet
  12. Place the filter trap assembly in the mounting bracket and install the retaining clip
  13. Reinstall the screws of the split vent valve
  14. Fully tighten the split vent front weldment onto the trap
  15. Turn on the column flow and inlet temperature and purge for 10 to 15 minutes
  16. Check for leaks, see: How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems 
  17. Install the pneumatics cover
  18. Reset the EMF counter if needed
  19. Restore the analytical method.

 Tip 
Tips: This procedure is also available in the video: Change the Split Vent Trap - GC Troubleshooting Series 

 Tip 
Learn how to effectively maintain your 7890 GC:
GC-7890-2232z - How to Replace the 7890 Split Vent Trap e-learning course available from Agilent education.
  • MMI
  • S/SL
  • GC Systems
  • Maintenance
  • PTV
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help