Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Inlet Liners for GC High Concentration Analyses
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • +GC Troubleshooting
    • -GC Best Practices
      • Capillary Column Installation Quick Reference Guide
      • Inlet Liners for GC High Concentration Analyses
      • Inlet Liners for GC Trace Analyses

Inlet Liners for GC High Concentration Analyses

Created by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet


Issue

Different liners are required for different high concentration, split analysis techniques including:

  • General use liners 
  • Auto inject, general use
  • Auto inject, viscous sample
  • Auto inject, dirty matrix
  • Manual slow injection
  • Headspace (HS) and purge and trap (P&T). 

Inlet liner choice is part of method validation, and the same liner type must be used consistently while running that method. For help with identifying the correct inlet liner for your application, see GC Liner Selector Tool.


Resolution

For the complete catalog of Agilent liners including multi value packs, click here.

  1. Split, general use
    Split taper, ultra inert (UI), glass wool, low pressure (LP), 870 μm, 4 mm ID p/n 5190-2295
    Single taper, universal UI frit, 870 μm, 4 mm ID p/n 5190-5105
        
  2. Auto inject, general use

    Single taper, universal UI glass frit, mid frit, 780 µL, 4 mm ID

    p/n 5190-5105
    Split taper, ultra inert (UI), glass wool, low pressure (LP), 870 μm,
    4 mm ID
    p/n 5190-2295
    Straight, UI, glass wool, 990 µL, 4 mm ID p/n 5190-2294
          
  3. Auto inject, viscous sample
    Split taper, ultra inert (UI), glass wool, low pressure (LP), 870 μm, 
    4 mm ID
    p/n 5190-2295
    Split taper, deactivated, glass wool, LP, 870 μm, 4 mm ID p/n 5183-4647      
               
  4. Auto inject, dirty matrix

    Direct Connect, single taper, glass wool, bottom hole, UI, 675 µL,
    4 mm ID

    p/n 5190-2295
    Split taper, deactivated, glass wool, LP, 870 μm, 4 mm ID p/n 5183-4647
    Single taper, deactivated, glass wool, 870 μm, 4 mm ID p/n 5183-4711
               
  5. Manual slow injection

    Dirty matrix

    Direct Connect, single taper, glass wool, bottom hole, UI, 675 µL,
    4 mm ID

    p/n 5190-2295
    Clean matrix
    Straight with cup, 870 μm, 4 mm ID p/n 18740-80190
          
  6. Headspace/purge and trap
    Straight, UI, 200 µL, 2 mm ID, for use with HS transfer line  p/n 5190-6168
    Straight, deactivated, quartz, 2 mm p/n 5181-8818
                
 Tip 
Learn how to effectively operate your Agilent GC System:
GC-0GEN-1012s - GC Inlets Theory and Operation e-learning course available from Agilent education

  • 6890
  • S/SL
  • 7820
  • Intuvo 9000
  • GC Systems
  • Best practices
  • 7890
  • 8890
  • 8860
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help