Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Peak Size Changes in GC
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • -Basic Troubleshooting for GC Systems
        • Source of Ghost and Carryover Peaks in GC
        • Excessive Background Noise in GC
        • Baseline Instability in GC
        • Peak Fronting in GC
        • Peak Tailing in GC Systems
        • Peak Splitting in GC
        • Peak Retention Time Shift in GC
        • Peak Size Changes in GC
        • Loss of Peak Resolution in GC
      • Using the Condensation Test to Troubleshoot the GC Inlet
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
      • Gathering GC Diagnostics Log Files Using the Agilent Acquisition Software
      • Gathering Diagnostics Log Files from the 8890/8860/Intuvo 9000 GC Systems Using the Browser Interface
    • +GC Best Practices

Peak Size Changes in GC

Created by Agilent Knowledge Portal Agilent Knowledge Portal 8 months ago | Last modified by Carlos Vargas Carlos Vargas 7 months ago

This Information Applies To: Agilent GC Systems


Issue

A change in peak size can occur for single, multiple or all peaks. 

Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems.


Background

Changes in peak size can be the result of changes analytical conditions, sample concentration, or injector problems (Figure 1):

           
Figure 1.  Change in peak size
1. Some peaks change size, 2. All peaks change size

Possible cause of change in peak size Solution
Coelution Improve separation of peaks, or clean up the sample
Change in sample concentration Check if new standards have been recently prepared. Changes may be caused by preparation error, degradation, evaporation, or variances in sample temperature or pH
Blocked or partially blocked syringe Check that the syringe plunger is moving freely. The solution should be pulled into the syringe body, and moving freely and fast when plunger is depressed
Leak in the syringe Replace the syringe. Leaks can occur around the plunger or around the needle; leaks are often not readily visible
Septum leak Replace the septum
Injector leak Leak check the injector: How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems 
Inlet contamination causing compound degradation  Replace the inlet liner, gold seal, bake out inlet Split/Splitless Inlet Bakeout Procedure 
Column contamination Bake out the column. Limit the bake-out for 1 to2 hours
 Caution 
Caution: Do not exceed the manufacturer recommended maximum column temperate limit as it will damage your column.
Trim the column: Remove 0.5 to 1 m from the front of the column. Trimming the column many times over its lifetime will change the effective column length, and impact retention times. Adjust instrument set column length every time the column is trimmed 
Column activity (active compounds) Irreversible. Replace the column
Change in injector discrimination (most severe for split injections) Check injector parameters
Inlet overload (vapor volume too large for installed inlet liner) Use Agilent Vapor Volume Calculator to adjust injection size, liner volume, inlet temperature, or solvent. Less injection volume and higher flow rates are helpful
Change in detector response (some or all peaks, and background) Check gas flows, temperatures, and settings
Change in background level or noise (peak area)  See Excessive Background Noise in GC  and Baseline Instability in GC 
Change in the method settings Check injection volume, purge activation time (splitless only), split ratio, septum purge flow, carrier gas flow, column temperature
Sample solvent incompatibility (splitless injections) Change the sample solvent or use a retention gap

 If these procedures have not resolved the change in peak size problems, contact Agilent Technical Support.

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1040z - Practical Steps in GC Troubleshooting
GC-0GEN-2040zs - GC Troubleshooting Series e-learning courses available from Agilent education
  • 7820
  • Intuvo 9000
  • GC Systems
  • Trobleshooting
  • 7890
  • 8890
  • 8860
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help