  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • -Basic Troubleshooting for GC Systems
        • Source of Ghost and Carryover Peaks in GC
        • Excessive Background Noise in GC
        • Baseline Instability in GC
        • Peak Fronting in GC
        • Peak Tailing in GC Systems
        • Peak Splitting in GC
        • Peak Retention Time Shift in GC
        • Peak Size Changes in GC
        • Loss of Peak Resolution in GC
      • Using the Condensation Test to Troubleshoot the GC Inlet
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
    • +GC Best Practices

Loss of Peak Resolution in GC

Created by Jo Boden Jo Boden 1 month ago | Last modified by Carlos Vargas Carlos Vargas 12 days ago

This Information Applies To: Agilent GC Systems


Issue

Loss of peak resolution can be a combination of decrease in separation and increased peak width. Loss of resolution can impact on analytical reproducibility and accuracy.

Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems.


Background

 Loss of resolution can be caused by decreasing separation between peaks or an increase in width of peaks (Figure 1):

           
Figure 1. Decrease in peak separation and peak broadening
1. Normal retention time separation, 2. Retention time difference, 3. Peak width at half height
4. Decrease in separation, 5. Increase in peak width

Possible cause of decrease in separation Solution
Change in column temperature (all peaks) Check the column temperature
Change in column dimension  Verify that the installed column matches the configured column in the instrument and the method column.
Trimming the column many times over its lifetime will change the effective column length, and impact retention times. Adjust instrument set column length every time the column is trimmed 
Coelution with another peak Improve peak separation. Change column temperature or stationary phase

Possible cause of increase in peak width Solution
Overloading from a large change in analyte concentration Dilute the sample
Change in sample concentration (higher concentrations have wider peaks) Check if new standards have been recently prepared. Changes may also be caused by preparation error, degradation, evaporation, or variances in sample temperature or pH
Column contamination Bake out the column. Bake-out for 1 to2 hours or until baseline is stable 
 Caution 
Caution: Do not exceed the manufacturer recommended maximum column temperate limit as it will damage your column.
Trim the column: Remove 0.5 to 1 m from the front of the column. Trimming the column many times over its lifetime will change the effective column length, and impact retention times. Adjust instrument set column length every time the column is trimmed 
Column activity (active compounds) Irreversible. Replace the column
Change in carrier gas flow Check the carrier gas flow rate
Change in the injector settings Check injection volume, purge activation time (splitless only), split ratio, septum purge flow, column temperature
Improper solvent effect, lack of focusing (splitless) Lower initial oven temperature, choose a different solvent, sample phase polarity match, use a retention gap

 If these procedures have not resolved the resolution problem, contact Agilent Technical Support.

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1040z - Practical Steps in GC Troubleshooting
GC-0GEN-2040zs - GC Troubleshooting Series e-learning courses available from Agilent education
  • Troubleshooting
  • 7820
  • Intuvo 9000
  • GC Systems
  • 7890
  • 8890
  • 8860
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help