Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Peak Retention Time Shift in GC
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • -Basic Troubleshooting for GC Systems
        • Source of Ghost and Carryover Peaks in GC
        • Excessive Background Noise in GC
        • Baseline Instability in GC
        • Peak Fronting in GC
        • Peak Tailing in GC Systems
        • Peak Splitting in GC
        • Peak Retention Time Shift in GC
        • Peak Size Changes in GC
        • Loss of Peak Resolution in GC
      • Using the Condensation Test to Troubleshoot the GC Inlet
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
    • +GC Best Practices

Peak Retention Time Shift in GC

Created by Jo Boden Jo Boden 1 month ago | Last modified by Carlos Vargas Carlos Vargas 8 days ago

This Information Applies To: Agilent GC Systems


Issue

A shift in retention time can occur for single, multiple or all peaks. 

Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems.


Background

Retention time (RT) shifts can be the result of changes analytical conditions, sample concentrations, column, or injector problems (Figure 1):

           
Figure 1. Retention time shift
1. Normal retention time, 2. Retention time increased (all analytes)

Possible cause of retention time shift Solution
Overloading from a large change in analyte concentration Dilute the sample
Septum leak Replace the septum
Injection technique Check that the syringe plunger is moving freely. The solution should be pulled into the syringe body, and moving freely and fast when plunger is depressed. Use an autosampler
Leak in the injector Leak check the injector How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems 
Change in carrier gas flow rate (all peaks similar RT shift) Check the carrier gas flow rate. Carrier gas flow rate can be measured with an unretained compound
Sample solvent incompatibility (splitless injections) Change the sample solvent or use a retention gap
Change in column temperature Check any changes in the method column temperature
Change in column dimension 

Verify that the installed column matches the configured column in the instrument and the method column.
Changes in column dimension impact carrier gas flow rate, a common cause is multiple trimming over the column lifetime
Trimming the column many times over its lifetime will change the effective column length, and impact retention times. Adjust instrument set column length every time the column is trimmed 


If these procedures have not resolved the retention time shift problem, contact Agilent Technical Support.

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1040z - Practical Steps in GC Troubleshooting
GC-0GEN-2040zs - GC Troubleshooting Series e-learning courses available from Agilent education
  • Troubleshooting
  • 7820
  • Intuvo 9000
  • GC Systems
  • 7890
  • 8890
  • 8860
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help