Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Baseline Instability in GC
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • -Basic Troubleshooting for GC Systems
        • Source of Ghost and Carryover Peaks in GC
        • Excessive Background Noise in GC
        • Baseline Instability in GC
        • Peak Fronting in GC
        • Peak Tailing in GC Systems
        • Peak Splitting in GC
        • Peak Retention Time Shift in GC
        • Peak Size Changes in GC
        • Loss of Peak Resolution in GC
      • Using the Condensation Test to Troubleshoot the GC Inlet
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
    • +GC Best Practices

Baseline Instability in GC

Created by Jo Boden Jo Boden 1 month ago | Last modified by Carlos Vargas Carlos Vargas 8 days ago

This Information Applies To: Agilent GC Systems


Issue

Baseline instability that is not reproducible across multiple runs can be introduced at different points including sample introduction, separation, and the detector. The baseline instability causes integration problems, which can lead to inaccurate quantitative or qualitative results.

Baseline rises reproducibly, are normal for ramped temperature programs. The change in baseline will be consistent across multiple runs and multiple days, and will not impact negatively on analytical results. 

Other chromatographic problems are identified in Basic Troubleshooting for GC Systems . Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems. 


Background

Over time sample deposits can accumulate in the analytical pathway and slowly leach out during subsequent sample analyses (see Figure 1) causing an inconsistently baseline across the run time. Baseline instability may or ay not be associated with high background noise: Excessive Background Noise in GC . 

   
Figure 1. Baseline instability
1. Normal background after baseline subtraction, 2. Baseline instability

Resolution

Possible sources of baseline instability include:

  1. Sample introduction system 

    Perform the AgilentGC Condensation Test. If the condensation test determines the sample introduction system is the source of the contamination, continue with the following:

    Possible source of contamination Solution
    Gas leak Test all gas lines for leaks. If the baseline instability coincided with a change in gas cylinder, start at the gas cylinder end of the gas lines
    Low quality or contaminated gas supply Check to see if the baseline instability coincided with a change in gas cylinder. Change the gas cylinder and flush the gas lines for sufficient time for the gas volume in the lines to be completely purged
    Septum bleed or degradation Replace the septum. Use an Agilent high-quality septum appropriate for your inlet temperature. See Replacing the Inlet Septum for Split/Spitless and Multimode Inlets 
    Inlet liner dirty Replace the inlet liner. Use an Agilent inlet liner appropriate for your injection volume, solvent, and injection mode
    Gold seal dirty Replace the gold seal. See How to Replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC 
    Inlet contamination Clean or Bake out the inlet. See How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems  


  2. Analytical column 

     Caution 
    Caution: Do not exceed the manufacturer recommended maximum column temperature limit as it will damage your column. 

    Possible source of contamination Solution
    Incompletely conditioned column Bake out the column. Limit the bake-out for 1 to 2 hours or until the baseline is stable
    Column contamination Bake out the column. Limit the bake-out for 1 to 2 hours or until the baseline is stable
    Trim the column: Remove 0.5 to 1 m from the front of the column

    Detector

    Possible source of contamination Solution
    Unequilibrated detector Allow the detector sufficient time to stabilize according to the user manual
    Detector leak Leak test the detector fittings

If these procedures have not resolved the baseline instability problem, contact Agilent Technical Support.

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1001e - Gas Management Best Practices
GC-0GEN-1012s - GC Inlets Theory and Operation e-learning courses available from Agilent education
  • Troubleshooting
  • 7820
  • Intuvo 9000
  • Detectors
  • GC Systems
  • 7890
  • 8890
  • 8860
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help