Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Basic Troubleshooting for GC Systems
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • -Basic Troubleshooting for GC Systems
        • Source of Ghost and Carryover Peaks in GC
        • Excessive Background Noise in GC
        • Baseline Instability in GC
        • Peak Fronting in GC
        • Peak Tailing in GC Systems
        • Peak Splitting in GC
        • Peak Retention Time Shift in GC
        • Peak Size Changes in GC
        • Loss of Peak Resolution in GC
      • Using the Condensation Test to Troubleshoot the GC Inlet
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
    • +GC Best Practices

Basic Troubleshooting for GC Systems

Created by Jo Boden Jo Boden 1 month ago | Last modified by Carlos Vargas Carlos Vargas 12 days ago

This Information Applies To: Agilent GC Systems


Issue

Several problems can be identified in GC chromatograms, and each of these problems can have various different sources.

Regular maintenance of your instrument in accordance with the recommended maintenance schedule will reduce the incidence of chromatography problems. 


Steps to follow

The most common gas chromatography problems identified in a chromatogram include: 

  • Ghost peaks or carryover (Figure 1). 
    See Source of Ghost and Carryover Peaks in GC 


    Figure 1. Ghost and carryover peaks
    1. Clean blank run (no injection), 2. Contaminated run with ghost peaks and carryover

  • Excessive baseline noise (Figure 2).
    See   Excessive Background Noise in GC 


    Figure 2. Excessive baseline noise
    1. Normal background with peaks visible, 2. Excessive background with peaks partially obscured

  • Baseline instability or disturbances (Figure 3).
    See Baseline Instability in GC 


    Figure 3. Baseline instability
    1. Stable baseline, 2. Unstable baseline

  • Fronting peaks (Figure 4).
    See  Peak Fronting in GC 

          
    Figure 4. Fronting peaks
    1. Normal symmetrical peak, 2. Fronting asymmetrical peak, 3. Overload

  • Tailing peaks (Figure 5) 
    See Peak Tailing in GC Systems 


    Figure 5.
    Tailing peaks
    1. Normal symmetrical peak

  • Split peaks (Figure 6).

    See Peak Splitting in GC 


    Figure 6. Split and broadening peaks

  • Retention time shift (Figure 7).
    See Peak Retention Time Shift in GC 


    Figure 7. Retention time shift
    1. Normal retention time, 2. Retention time increased

  • Change in peak size (Figure 8). 
    See Peak Size Changes in GC 

       
    Figure 8. Change in peak size
    1. Some peaks change size, 2. All peaks change size

  • Loss of resolution (Figure 9). 
    See Loss of Peak Resolution in GC 


    Figure 9. Loss of separation
    1. Normal retention time separation, 2. Separation, 3. Peak width
    4. Decrease in separation, 5. Increase in peak width

 Tip 
Learn how to effectively troubleshoot your Agilent GC System:
GC-0GEN-1040z - Practical Steps in GC Troubleshooting
GC-0GEN-2040zs - GC Troubleshooting Series e-learning courses available from Agilent education
  • Troubleshooting
  • 7820
  • Intuvo 9000
  • Detectors
  • GC Systems
  • 7890
  • 8890
  • 8860
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help