GC Articles Agilent GC Detector Sampling (Data) Rate
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • -GC Operation
      • How to Use the Browser Interface for 8890 and 8860 GC Systems
      • Cleaning Local Data Storage From 8890 or 8860 GC Systems
      • How to Lock/Unlock GC Columns with Agilent OpenLab CDS
      • Agilent GC Detector Sampling (Data) Rate
      • Configuring the Flame Ionization Detector Makeup Gas for Agilent 7890 GC Systems
    • +GC Maintenance
    • +GC Troubleshooting
    • +GC Best Practices

Agilent GC Detector Sampling (Data) Rate

Created by Agilent Knowledge Portal Agilent Knowledge Portal 9 months ago | Last modified by Carlos Vargas Carlos Vargas 7 days ago

This Information Applies To: Agilent GC Systems


Issue

The data rate determines the number of sampling points across the chromatographic peak and the sample background. The data rate is based on the width of the analyte speak (in minutes) and the ratio of the analyte signal to the background noise. 


Background

The data rate unit is hertz (Hz), where 1 Hz is one cycle per second. The data rate determines the number of data points taken across the width of the peak, and in the background. The ratio of the signal peak height to the background peak height determines the detection limit. For more information on calculating and checking the signal-to-noise see How to Check the signal-to-noise ratio with Agilent OpenLab CDS (version 2.x) .

  • Faster data rates (higher Hz):
    • Takes more readings per second, increasing the peak height, peak symmetry, noise, and the sample data file size. 
    • Improves signal-to-noise
  • Slower data rates:
    • Decrease background noise, peak height and peak symmetry, the noise ratio and the data file size. 
    • Decreases signal-to-noise.

The impact on signal-to-noise ratio (relative noise level) is demonstrated in (Figure 1):


Figure 1. Impact on data rates on signal-to-noise
1 signal-to-noise, 2 Faster data rates, 3 Slower data rates
 - - -  signal-to-noise, — Excess noise due to flow, oven temperature, detector temperature 

Agilent recommends 10 to 20 data points across a peak to optimize peak height and signal-to-noise. Use the guidelines in Table 1 to assist selecting the data rate. 

Data rate

(Hz)

Minimum peak width

(minutes)

Number of data points
Signal-to-noise Detector Column Type
500 0.0001 3  5 FID Narrow-bore. 0.05 mm 
200 0.001 12 3.1 FID Narrow-bore. 0.05 mm
100 0.002 12 2.2 FID/FPD/NPD only

Capillary

to

packed

50 0.004 12 1.6 All types
20 0.01 12 1
10 0.02 12 0.7
5 0.04 12 0.5
2 0.1 12 0.3 All types excluding the TCD*
1 0.2 12 0.22
0.5 0.4 12 0.16
0.2 1.0 12 0.10 
0.1 2.0 12 0.07

Table 1. Date rate guidelines.

* The thermal conductivity detector (TCD) switching rate is 5 Hz. A setting of less than 5 Hz can cause tail ringing (Figure 2):


Figure 2. TCD tail ringing

 Tip 
Learn how to effectively operate your Agilent GC System:
GC-0GEN-1013s - GC Detectors Theory and Operation
GC-0GEN-1051r - FID Theory and Operation e-learning courses available from Agilent education
  • Operation
  • FPD
  • Detectors
  • GC Systems
  • TCD
  • uECD
  • NPD
  • FID
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help