Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Gas Chromatography Portal
  • Agilent Knowledge Portal
Gas Chromatography Portal
GC Articles Agilent Thermal Conductivity Detector Troubleshooting
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • +GC Maintenance
    • -GC Troubleshooting
      • How To Ping the IP Address of an Agilent Gas Chromatographs
      • Checking for External Leaks on 7697A Headspace
      • Troubleshooting a Leak or Poor Chromatography on the Agilent 7697A Headspace Sampler
      • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint
      • Troubleshooting a Leak in a Split/Splitless Inlet
      • How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
      • How to Bake-out the Split/Splitless Inlet (S/SL) on Agilent GC Systems
      • Troubleshooting an FID not Igniting
      • Agilent GC FID Jet Restriction Test
      • Troubleshooting FID Noise and High Background
      • Agilent Thermal Conductivity Detector Troubleshooting
      • ECD Baseline Troubleshooting
    • +GC Best Practices

Agilent Thermal Conductivity Detector Troubleshooting

Created by Jo Boden Jo Boden 1 month ago | Last modified by Carlos Vargas Carlos Vargas 23 days ago

This Information Applies To: Agilent 8890/8860/7890/7820/6850/6890 GC Systems with TCD (Thermal Conductivity Detector)


Issue

Problems observed with thermal conductivity detector (TCD) results can have different underlying causes. 


Resolution

Check that the TCD settings are within the recommended range, then assess the instrument for the following errors/problems:

Hardware Fault Error No signal and/or the GC front panel screen displays Front/Back Detector Error.

Cause: Filament is open (burnt out)
Fix: Replace the TCD detector cell. Contact Agilent Technical Support.

High background causing poor sensitivity

Cause: Contamination of the gas flow system
Fix:

  1. Check gases are of 99.999% purity / 5.0 grade
  2. Check column is installed correctly (see Capillary Column Installation Quick Reference Guide )
  3. Leak test the inlet (see Troubleshooting a Leak in a Split/Splitless Inlet  and How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems )
  4. Leak test the TCD
  5. Bake out the TCD 

Drifting or wandering baseline

Cause: Some drift is normal in temperature programmed analyses, but should be reproducible between runs. To determine if the drift is greater than normal compare with established baseline.
Fix:

  1. Check heaters/ sensors to verify there's no active faults: Power cycle the instrument to start a self-check of all thermal heated zones. Any detected faults will be reported on screen
  2. Leak test the inlet (see Troubleshooting a Leak in a Split/Splitless Inlet  and How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems )
  3. Leak test the TCD
  4. Bake out the TCD 

Ringing on Peak Tail



Figure 1. Ringing at peak tail

Cause: The sampling rate for the detector is set to equal to or less than 5 Hz (which is the switching rate of the solenoid valve). For more information on sampling rate see Agilent GC Detector Sampling (Data) Rate 
Fix: Change the sampling rate to greater then 5 Hz. 

 Tip 
Tip: Calibration curves must be re-established after any changes to the sampling rate. In some cases, methods may require re-validation. 


Noise/Spikes in Baseline

Causes include: 

  1. A Valve (Sampling or Switching types) in the GC column flow path switching causing a pressure / flow disturbance
  2. Power problems, including poorly grounded power
  3. Inconsistent gas quality, pressure, or flow being delivered to the GC, gas generators are more prone to supply problems that gas cylinder supplied gas
  4. Rapid ambient pressure changes in the area of the GC, e.g., air-conditioning fans cycling on/off or opening a door between areas with an intentional pressure difference.
  5. Faulty electronics board.

Fix: Contact Agilent Technical Support for further assistance.

 Tip 
Learn how to effectively operate and troubleshoot your Thermal Conductivity Detector (TCD):
GC-0GEN-1013s - GC Detectors Theory and Operation
GC-7890-2281r - Agilent 7890 Series GC Thermal Conductivity Detector (TCD) Troubleshooting e-learning courses available from Agilent education
  • Troubleshooting
  • Detectors
  • GC Systems
  • TCD
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help