GC Articles How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • -GC Maintenance
      • -GC Systems
        • Replacing the Inlet Septum for Split/Spitless and Multimode Inlets
        • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC
        • Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System
        • Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System
        • How to Clean an FID
        • Bake Out the Agilent Flame Ionization Detector
        • Bake Out the Agilent Thermal Conductivity Detector
        • Bake Out the Agilent Micro Electron Capture Detector
        • Agilent GC Firmware Update Procedure
        • How To Find More Information About the Agilent Gas Chromatograph micro-ECD Wipe Test Kit
        • How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 GC Systems
        • How To Clear the Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatograph
        • Column Configuration for the 8860/8890 GC System
        • Column Configuration for the 7890 GC System
      • +GC Samplers
    • +GC Troubleshooting
    • +GC Best Practices

How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC

Created by Anna Gärtner Anna Gärtner over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 5 months ago

This Information Applies To: Agilent 6890, 7890, 7820, 8890 and 8860 GC Systems with Split/Splitless (S/SL) inlet


Issue

Replacement of the Gold Seal is required as part of inlet maintenance. Replacement is required when the performance of the inlet degrades over time due to accumulated contamination (e.g., baseline noise, ghost / artifact peaks, loss of reproducibility, carry over, loss of compound response or no compound response).


Items Required

The following Gold seals are available

  • Ultra Inert Gold Seals (p/n 5190–6144) can be used for inert applications.
  • Standard Gold Seals (p/n 5188–5367) can be used regular applications.
  • Cross-Cut Gold Seal (p/n 5182–9652) should be used when the S/SL Inlet total flow is set ≥ 200 ml/min.

Steps to follow

 Tip 
Tips: Wear clean, lint-free gloves to prevent contamination of parts with dirt and skin oils.

 Tip 
Tips: The 8890/60 GC Systems offer built-in guided maintenance procedures that follow the procedures described in this article.
Go to Maintenance > (Front/Back) Inlet > Perform Maintenance. Select Replace Gold Seal and then "Start Maintenance".
This feature resets the EMF counters automatically when completed.


Figure 1 Guided Maintenance procedure for 8890/60 GC Systems

  1. Cool the GC Oven (set to 35 °C), set Inlet Temperature Off, Detector Temperature can remain unaltered. 

     Warning  Warning: The oven, inlet, and/or detector may be hot enough to cause burns. If the oven, inlet, or detector is hot, wear heat-resistant gloves to protect your hands.

     Caution  Caution: If using a TCD the filament in the TCD will be damaged if exposed to air while hot. To protect the filament, turn it Off.

  2. Set the inlet pressure to 0.0 psi.

     Caution  Caution: Wait for the inlet, oven, and column to cool down to near room temperature to protect the column before turning off inlet / column carrier gas flow from the GC control software or GC front panel. 

  3. Remove the inlet liner using tweezers.


    Figure 2 Removing the liner and O-ring

  4. Remove the column from the inlet. Cap the open end of the column with a septum to prevent contamination and air entering the column. 


    Figure 3 Capping the capillary column with a septum.
    1. Insulation cup, 2. Septum.

  5. Remove the insulation cup and insulation around the base of the inlet by loosening the two screws holding it and then using a clockwise-twisting motion.


    Figure 4 SSL inlet insulation cup.

  6. Loosen and remove the reducing nut. Remove the washer and gold seal inside the reducing nut.


    Figure 5 Removal of reducing nut.

  7. Put a new washer in the reducing nut and place the new gold seal on top of it (flat side facing up).


    Figure 6 1. Gold seal, 2. Washer, 3. Reducing nut, 4. This side faces down.

  8. Reinstall the reducing nut and tighten securely with a wrench. The gold seal  is retained by the reducing nut, which must be tightened sufficiently to make this seal gas tight.

  9. Replace the inlet liner if necessary.

  10. Reinstall the insulation cup and the column.

  11. Reset the inlet pressure and allow the inlet and column to purge with carrier gas for at least 15 minutes before heating the inlet and / or the column oven.

  12. Check for leaks by performing the Leak and Restriction Test.

  13. Reset the EMF counter if needed.

     Tip 
    Tips: See "To Reset an EMF Counter" (7890 GC) or "To Enable, Reset, or Change a Limit for an EMF Counter" (8890/60 GCs) in the respective Operation Manual.

  14. Bakeout contaminants. 

     Tip 
    Tips: See "To Bakeout Contaminants from the Split/Splitless Inlet" in the respective Maintenance Manual.

  15. Restore the analytical method.

     Tip 
    Tips: This video will guide you through the gold seal replacement procedure.

 Tip 
Learn how to effectively maintain your 7890 and 8890 GC:
GC-7890-2240s - Agilent 7890 Series GC: Preparing for Maintenance and Maintenance Utilities
GC-8890-2200s - Agilent 8890 GC Maintenance and Troubleshooting e-learning available from Agilent education.

  • 6890
  • S/SL
  • GC Systems
  • Maintenance
  • 7890
  • 6850
  • 8890
  • 8860
  • Inlets
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help