GC Articles How To Clear the Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatograph
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • -GC Maintenance
      • -GC Systems
        • Replacing the Inlet Septum for Split/Spitless and Multimode Inlets
        • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC
        • Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System
        • Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System
        • How to Clean an FID
        • Bake Out the Agilent Flame Ionization Detector
        • Bake Out the Agilent Thermal Conductivity Detector
        • Bake Out the Agilent Micro Electron Capture Detector
        • Agilent GC Firmware Update Procedure
        • How To Find More Information About the Agilent Gas Chromatograph micro-ECD Wipe Test Kit
        • How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 GC Systems
        • How To Clear the Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatograph
        • Column Configuration for the 8860/8890 GC System
        • Column Configuration for the 7890 GC System
      • +GC Samplers
    • +GC Troubleshooting
    • +GC Best Practices

How To Clear the Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatograph

Created by Vincent Nguyen Vincent Nguyen over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 9 days ago

This Information Applies To: Agilent 8890/8860 GC Systems


Issue:

This procedure outlines the steps to view and clear the maintenance warnings on based on Agilent 8890/8860 GCs. You can also follow the onscreen instructions to perform maintenance steps, then reset the EMF counters for that particular maintenance item.


 Tip 
Note: For instructions on how to set up the EMF, see the article on How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatographs.

Steps to follow

1. Notice the warning exclamation mark (!) on the Maintenance ribbon, which indicates that you have one or more past due EMF.

Note: There is either yellow or red warming for either past "Service Warning" or "Service Due," respectively.

Agilent - EMF Maintenance Inlet Warning.png Agilent - EMF Maintenance Inlet Service Due.png

2. Click on the "Maintenance" ribbon on the top of the ribbon pan of the GC to view the past due items. For illustration, the GC inlet in this case the inlet liner age has exceeded its EMF set point. 

Agilent - EMF Maintenance Inlet Warning.png

Agilent - EMF Maintenance Inlet Warning Inlet Settings.png

3. If the liner age is past the service warning – the display will show a yellow warning badge in the ‘Inlets Maintenance’ screen, the ‘Maintenance’ screen, and on the ‘Maintenance’ tab on the top ribbon.

4. If  the ‘Service Due’ Threshold is exceeded, the warning will turn red. In this case the service due is at three weeks:

Agilent - EMF Maintenance Inlet Service Due.png

Agilent - EMF Maintenance Inlet Service Due Items.png

Agilent - EMF Maintenance Inlet Service Due Inlet Settings.png

5. Clear the ‘Service Warning’ or the ‘Service Due’ badge by completing the maintenance task or by manually resetting the counter:

a. Complete the maintenance task:

1. To Complete the Maintenance task: select "Perform Maintenance" on the inlet.

Agilent - EMF Start Maintenance.png

2. Select "Replace Liner" then click on " Start Maintenance." The display will show a step-by-step guide of replacing the liner. Once all the steps have been completed, the Early Maintenance Feedback counter will automatically reset. 

Agilent - EMF Start Maintenance Replace Liner.png

Agilent - EMF Start Maintenance Replace Liner Tools.png

b. Manually Reset EMF:

1.To Manually reset the Early Maintenance Feedback, select the ‘Liner Age’ from the ‘Inlets Maintenance’ screen to get back to the ‘Liner Age Settings’ window and select ‘Reset Counter’

Agilent - EMF Maintenance Manual Reset.png

2. Select ‘yes’ when getting an ‘Are you sure…’ message. This action will clear the ‘Service Due’ or ‘Service Warning’ badge:

Agilent - EMF Maintenance Manual Reset Confirm.png

Agilent - EMF Maintenance Manual Reset All Clear.png

 Bulb 
Learn how to effectively maintain your Agilent 8890/8860 gas Chromatographs:
Agilent 8860 GC with OpenLab CDS ChemStation Edition 
Agilent 8860 GC with OpenLab CDS 
Agilent 8890 GC with OpenLab CDS ChemStation Edition 
Agilent 8890 GC with OpenLab CDS e-learning paths available from Agilent education
  • Troubleshooting
  • GC Systems
  • Maintenance
  • 8890
  • 8860
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help