GC Articles How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 GC Systems
  • GC Articles
  • GC Videos
  • GC Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Gas Chromatography
    • +GC Operation
    • -GC Maintenance
      • -GC Systems
        • Replacing the Inlet Septum for Split/Spitless and Multimode Inlets
        • How to replace the Gold Seal on the Split/Splitless (S/SL) Inlet for an Agilent GC
        • Replacing the Split Vent Trap Cartridge Filter on a 8890/8860 GC System
        • Replacing the Split Vent Trap Cartridge Filter on a 7890/7820 GC System
        • How to Clean an FID
        • Bake Out the Agilent Flame Ionization Detector
        • Bake Out the Agilent Thermal Conductivity Detector
        • Bake Out the Agilent Micro Electron Capture Detector
        • Agilent GC Firmware Update Procedure
        • How To Find More Information About the Agilent Gas Chromatograph micro-ECD Wipe Test Kit
        • How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 GC Systems
        • How To Clear the Early Maintenance Feedback (EMF) on the Agilent 8890/8860 Gas Chromatograph
        • Column Configuration for the 8860/8890 GC System
        • Column Configuration for the 7890 GC System
      • +GC Samplers
    • +GC Troubleshooting
    • +GC Best Practices

How To Use and Configure Early Maintenance Feedback (EMF) on the Agilent 8890/8860 GC Systems

Created by Vincent Nguyen Vincent Nguyen over 1 year ago | Last modified by Carlos Vargas Carlos Vargas 4 months ago

This Information Applies To: Agilent 8890/8860 GC Systems


Issue

Early Maintenance Feedback is a useful feature on Agilent GC Systems and software that allows a user to monitor the maintenance needed on their GC. This feature allows for users to get a reminder that maintenance is due on their instrument before it causes a chromatography or even a hardware issue. If you are changing parts on your GC on a scheduled basis, either based on a certain passing of time or a specific number of injections, Early Maintenance Feedback (EMF) is the tool for you. 


 Tip 
Note: For instructions on how to set up the EMF, see the article How To Clear Early Maintenance Feedbacks (EMFs) on the Agilent 8890/8860 Gas Chromatograph

Steps to follow 

1. Press Maintenance Agilent-Maintenance Ribbon.png on the top ribbon pane on the 8890 GC screen.

Agilent-8890 Touch Screen Display.png

2. You will see the components of your GC on the Maintenance screen. In this case, we have ALS, Inlets, Columns, Detectors, and Instrument.

Agilent - EMF Maintenance Screen.png

3. For example, if you press Inlets, it will take you to the Inlets Maintenance screen. You will be able to see both inlets from the tabs on top. In the picture below, we are looking at the Split-Splitless (SS) inlet. On the second tab, there is a Purge Packed (PP) inlet on the back of this GC.

Agilent - EMF Inlets.png

4. From here you can select gold seal age, gold seal injections, liner age, liner injections, liner O-ring age, liner O-ring injections, septum injections, and split vent trap age, and split vent trap injections. You can press the down-facing arrow to bring up more option and information.

5. If you press on any of these options, a dialogue box will open with your EMF Settings for that specific item. For example, if you press Liner Age, you will see the Liner Age Settings dialogue box below: 

Agilent - EMF Inlets Settings.png

6. Here you can set the Early Maintenance Feedback for the Liner Age. You can set a Service Warning or a Service Due. You can choose the time frame by pressing n the drop-down menu to select the time unit for your EMF setting, such as weeks, years, days, hours, minutes, or seconds. Once you have selected the time unit, you can click the field to the left of the unit to edit your time interval numeric value.

7. You can then can check the enable box and then apply to set your EMF for that particular item.

 Bulb 
Tips: You can also access the EMF using a Web Browser Interface, as shown in the following video. 

 Bulb 
Learn how to effectively maintain your Agilent 8890/8860 gas chromatographs:
Agilent 8860 GC with OpenLab CDS ChemStation Edition
Agilent 8860 GC with OpenLab CDS
Agilent 8890 GC with OpenLab CDS ChemStation Edition
Agilent 8890 GC with OpenLab CDS e-learning paths available from Agilent education
  • Troubleshooting
  • GC Systems
  • Maintenance
  • 8890
  • 8860
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help