Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • GC
    • GC/MS
    • LC
    • LC/MS
    • Life Sciences Reagents
    • Molecular Spectroscopy
      Molecular Spectroscopy
      • Molecular Spectroscopy - UV-Vis & UV-Vis-NIR, FTIR, Raman & Fluorescence
      • 8700 LDIR User Group
      • Transmission Raman User Group
    • Real-Time PCR (qPCR)
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
GC Portal
  • Agilent Knowledge Portal
GC Portal
Articles How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems
  • Articles
  • Videos
  • Forum
  • Announcements
  • More
  • Cancel
  • Agilent Knowledge Portal
  • GC Portal Home
  • GC Portal Articles
  • GC Portal Videos
  • More
  • Cancel
  • Created :  10 Nov 2022
  • Modified :  14 Sep 2023
  • Category :  Agilent Knowledge Portal
  • Entry Type :  Article
  • Product Type :  GC Systems
  • Component :  GC Inlets
  • Product Name :  7890 GC System 8860 GC System 8890 GC System
  • Task :  Maintenance Troubleshooting
Related Articles
  • Capillary Column Installation Quick Reference Guide

    This Information Applies To: All Agilent Gas Chromatographs, 6890, 6850, 7890, 7820, 8890, and 8860. Issue: A quick reference guide on the capillary column installation for the GC. Inlets Detectors Split/splitless inlet Purged packed inlet Multimode Inlet Cool on Column (COC) Inlet PTV Inlet Volatiles Interface (VI) Inlet FID / NPD TCD - Standard Low Leakage Ferrules TCD - Original Design Column Adapter µECD FPD Steps to follow Warning : The...
  • Replacing the Inlet Septum for Split/Spitless and Multimode Inlets

    This Information Applies To: Agilent Intuvo 9000, 8860, 8890, 7890A/B, 7820A, 6890, 6850 GC Systems with Split/Splitless, or Multimode Inlets. Issue: This article describes how to replace the inlet septum for Split/Splitless and multimode inlets for all Agilent GC Systems. Regular replacement of the septum can reduce small leaks in the inlet, allow for more reproducible chromatography, and protect the column from degradation. Replacement intervals should be every 100 to 400 injections depending...
  • Inlet Liners for GC Trace Analyses

    This Information Applies To: Agilent GC Systems with Split/Splitless (S/SL) inlet Issue Different liners are required for different splitless trace analysis techniques including: General use liners Pressure, temperature, volume (PTV) Hot inlet High recovery Cold multimode Headspace (HS) and purge and trap (P&T) Solid phase micro extraction (SPME) . Inlet liner choice is part of method validation, and the same liner type must be used consistently while running that method...
  • Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint

    This Information Applies To: Agilent GC systems, focusing on 8890 and 7890A/B, but also applies to 6850A/B, 6890A/N, 7820A, and 8860. Issue: If the inlet pressure / carrier gas flow fail to achieve the setpoint after a few minutes, the inlet pressure will shut down automatically. This article will help you to identify the possible causes of this issue. Steps to follow: Leak at the inlet To detect big leaks in the inlet, perform one of the diagnostic tests for the inlet: Inlet Leak Check...
  • Troubleshooting a Leak in a Split/Splitless Inlet

    This Information Applies To: Agilent GC Systems with Split/Splitless Inlet. Issue A leak may produce the following symptoms: Inlet pressure unable to reach setpoint (probably a large leak) - will lead to a flow shut-down. Failure of Any Diagnostic Tests from the associated tab on the front panel of the GC and/or the Browser Interface Leak & Restriction Test Pressure Decay Test Septum Purge Test Split Vent Restriction Test The cause of many chromatography issues can be leak...

How to Run and Interpret the Inlet Leak and Restriction Test for the Agilent 7890/8860/8890 GC Systems

Answer

This Information Applies To: Agilent 7890 A/B, 8860 and 8890 GC Systems with Split/Splitless (SSL), Multimode (MMI) or Programmable Temperature Vaporization (PTV) Inlets


Issue:

The Inlet Leak & Restriction Test is executed after every inlet maintenance procedure. It can also be run for troubleshooting purposes or as a frequent confidence check for GC performance.

Useful to

  • Find and repair medium to large leaks. Useful when the inlet will not pressurize to the setpoint due to a large leak: Troubleshooting a Leak in a Split/Splitless Inlet , Inlet Pressure Shutdown / Gas Cannot Achieve Setpoint .
  • Confirm that the column is not restricted and actually has carrier flow.
  • Check day to day leak rate. Run every morning before running samples. Confidence check. (Can be executed as a Prerun test as part of the method.)

Steps to follow:

Caution
Caution: The accuracy of the Leak and Restriction Test is directly related to the accuracy of the column configuration and carrier gas configuration. Confirm the configuration is correct prior to performing the test!


How-to perform an Inlet Leak and Restriction test on a 7890

  1. On the keypad press: [Service Mode] > Inlet Leak Check and confirm with [Enter]. (Figure 1)

  2. Set a test pressure. Usually, the default 10 psi pressure is suitable for the test, but it actually depend on the column dimensions, see the following examples:

    1. For 30 m X 250 µm on GC/MS: 5–20 psi.

    2. For 30 m X 530 µM column on GC Detector: 5 to15 psi.

    3. For 30 m X 320 µM column on GC Detector: 5 to 25 psi.

    4. For 30 m X 200 µM column on GC Detector: 5 to 85 psi.

       Caution 
      Caution: With a GC/MS system, use a test pressure that will not result in excessive flow to the MS.

       

  3. Set Warning or Fault thresholds if needed. (Figure 2) More information is available in the Troubleshooting Manual.

    1. A warning will create a "Service Due" reminder.

    2. A fault will set the GC in a "Not Ready" state.

       Caution 
      Caution: On a 7890 GC the leak check will always report "OK" unless a Warning or Fault Threshold was set prior.

       

  4. Press [ON] to start the Leak Check.
     
Figure 1 Selecting the Leak and Restriction Test on a 7890.

 

Figure 2 Setting of Warning and Fault thresholds.

How-to perform a Leak and Restriction test on a 8890/8860

  1. Select the automated test

    1. on the 8890 touch screen display press: Diagnostics > Inlet > Leak & Restriction Test. (Figure 3) 

    2. on the 8890 or 8860 Browser Interface (How to Use the Browser Interface for 8890 and 8860 GC Systems ): Diagnostics > Diagnostic Tests > Inlet > Leak & Restriction Test. (Figure 4)

  2. Press Start Test. 

    1. The test shuts off all exits to flow in the inlet system except the column. The target flow is set as 4 ml/min ± 2 ml/min. The diagnostic measures the column flow and calculates the Leak Rate.

    2.  During the test details are displayed along with test results. (Figure 5)
        

Figure 3 Selecting Leak and Restriction Test on the 8890 GC touch screen display. Figure 4 Selecting Leak and Restriction Test through the 8890 or 8860 Browser Interface.

 

Figure 5 Leak and Restriction Test page on the 8890 touch screen display.


How-to interpret the results of a Leak and Restriction Test

Tip With the Leak and Restriction Test leaks > 2 ml/minute can be detected.


If the actual is the target
 ± 2 ml/minute, the test is considered passed.

  • If you suspect smaller leaks, perform a Pressure Decay Test. (Troubleshooting Manuals)

If the actual is larger than the target, a leak exists in the flow path.

  • To identify a leak, check possible leak sources Troubleshooting a Leak in a Split/Splitless Inlet . If available, use an electronic leak detector.

  • Remove any leak sources from the previous step.  Run the Leak and Restriction test to confirm “Passed” status.

  • If the leak can't be identified, contact your local Agilent Technical Support.

If the actual is less than the target, a restriction exists in the flow path.

  • Check if a clean and correct liner is installed, the gold seal is not contaminated / blocked, and the column is not blocked.

  • If the restriction can't be identified, contact your local Agilent Technical Support.

Actions to take on a 7890

 Caution 
Caution: On a 7890 GC the leak check will report "OK" unless a Warning or Fault Threshold was set prior. Check the Total Flow and Column Flow results of the test.

 

  1. Two flow values are presented on the GC display. The approximate Leak Rate is the difference between the actual measured flow ("Total Flow") and the calculated column flow ("Col"). (Figure 6.1, Figure 6.2)

  2. If the test fails, touch or tighten each of your potential leak points while watching the display. When you touch the primary leak point, the “Total Flow” will respond up or down in the display.  

  3. Reset Early Maintenance Feedback counters (EMF) if applicable after change of inlet consumables. More information is available in the respective Operation Manual.

  

Figure 6.1 Passed Leak and Restriction Test. Figure 6.2 Failed Leak and Restriction test.


Actions to take on a 8860/8890

 Bulb  The Active Monitor gives the user the ability to troubleshoot the problem live during the test.

 

  1. If the test fails, select Active Monitor to isolate the leak(s) while the pressure is displayed. (Figure 7). Chec possible leak points while measuring the pressure on the screen.

  2. Reset Early Maintenance Feedback counters (EMF) if applicable after change of inlet consumables. More information is available in the respective Operation Manual

  3. If the leak can't be identified, contact your local Agilent Technical Support..

 

Figure 7 Active Monitor option. In this example, the failure is a leak.

 

 Bulb 
Learn more about how to troubleshoot your Agilent GC System:
Agilent 8890 GC with OpenLab CDS
Agilent 7890 Series GC with OpenLab CDS e-learning paths available from Agilent education.

 

 