This information applies to: Agilent OpenLab ChemStation Edition C.01.08 or later – Standalone Workstation.
Issue
When a PC name has changed, OpenLab ChemStation will no longer launch the configured instrument due to a license error or instrument controller error.
Background
Steps to Follow
Ensure that authentication is turned off by going to Control Panel of OpenLab ChemStation > Administration > Local Configuration > Authentication – None. If the authentication is set up, log in as an administrator and turn off authentication.
Caution: Before setting the authentication provider to None, it is highly recommended to create a backup of the Shared Services using OpenLab Shared Services Maintenance so none of the user group settings are lost.
|
There are three solutions depending on the user situation:
Solution 1: User can change the current PC name back to the original PC name.
If possible, revert the PC name to the previous name permanently to match the original instrument controller. The original PC name can be found in the Control Panel of OpenLab ChemStation > Administration > Instrument Controller > PC Name (Figure 1).
As the specific steps to change PC names can change between different versions of Windows OS, it is best to do an Internet search for a method to change your PC name that works with the specific version of Windows OS you are currently running on your Workstation PC.
Figure 1. PC Name in OpenLab Control Panel
Solution 2: User can temporarily change PC name back to the original PC name.
Figure 2. Backup ChemStation.ini file
1. Hidden Items, 2. ChemSation.ini file
Figure 6. Old and New Instrument Controllers
1. Old Instrument Controller, 2. New Instrument Controller
Solution 3: Cannot change PC name at all.
Tip: If creating a new instrument does not resolve this issue, follow the official uninstallation and reinstallation instructions to reinstall the OpenLab ChemStation software.
Full installation instructions can be found in the Workstation Installation and Configuration Guide for the specific version of OpenLab CDS ChemStation you are using to control your instrument. Search on Agilent.com for further information. |