Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG

Created by Jasmin Manck-Goetzenberger Jasmin Manck-Goetzenberger over 1 year ago | Last modified by Geoff Pettitt Geoff Pettitt 4 months ago

This Information Applies To: Fragment Analyzer, Femto Pulse, ZAG


Issue:

A new syringe is required to clean the reservoir vent valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG (see Figure 1).

Figure 1. Syringe (marked in red square) mounted to the reservoir vent valve of a Fragment Analyzer.

For all instruments, perform reservoir vent valve cleaning when needed, for example when experiencing migration issues, or dried gel is visible in the valve. Exchange the syringe whenever it wears out or is dirty. For a detailed description of the reservoir vent valve cleaning process for your instrument, please check the indicated pages in the respective manual:

  • Fragment Analyzer system manual, page 63
  • Femto Pulse system manual, page 61
  • ZAG system manual, page 56


Resolution: 

Unfortunately, Agilent Technologies does not offer the disposable syringe, but you can search online for 10 mL Luer Lock tip plastic syringe. The Luer lock tip is compatible to the reservoir vent valve.

 Bulb 
Learn how to effectively maintain your Agilent parallel capillary electrophoresis instruments:
Fragment Analyzer system manual
Femto Pulse system manual
ZAG system manual


For Research Use Only. Not for use in diagnostic procedures.


PR7000-7924

  • ZAG DNA Analyzer
  • Troubleshooting
  • Femto Pulse
  • Maintenance
  • Parallel Electrophoresis
  • Fragment Analyzer Systems
  • Fragment Analyzer
  • Femto Pulse System
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help