Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments

Created by Carlos Vargas Carlos Vargas 11 months ago | Last modified by Shweta.S Shweta.S 1 month ago

This Information Applies To: 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, Oligo Pro II


Issue:

This article explains the steps to take after installing a brand new array in the Agilent 5200/5300/5400 Fragment Analyzer, Femto Pulse, ZAG, and Oligo Pro II systems.


Steps to follow

  • Capillary array replacement steps are detailed in chapter 8 of the Fragment Analyzer user manual, Femto Pulse user manual, ZAG DNA analyzer user manual, and chapter 9 of the Oligo Pro II user manual.
  • After the new capillary array has been installed, perform the following steps in the order in which they are listed. 

1. Update the Array Serial Number - After the installation of the new capillary array, log in as a user with administrator privileges. To update the serial number in the instrument controller software, click Admin > Configuration (Figure 1). The Configuration Settings dialog box will then appear. Select Device Settings, enter the new serial number, and change the information in the drop-down lists on the Device Settings tab as appropriate (Figure 2).

            

Figure 1. Configuration command from controller software.  

 

Figure 2. Update device settings and enter new array serial number. 

2. Conditioning method run - For the Fragment Analyzer, Femto Pulse, and ZAG DNA Analyzer, run a Method C Flush to ensure optimal performance. Figure 3 shows screenshot from Fragment Analyzer controller software with Method C selected. Please note for Oligo Pro II, it is recommended to do a conditioning run. There is no Method C method file on the Oligo Pro II by default.

   

Figure 3. Conditioning method setup

3. Normal separation run – Next, a separation can be run either with samples, or with each well containing diluent marker or blank solution only.

4. Capillary array alignment - Once the run is finished, in the controller software, click Utilities> Capillary Alignment and perform a capillary alignment according to steps listed in the instruments' respective user manual. Briefly, there are two ways to perform a capillary alignment, though Method A is the recommended and easiest method of alignment. Figure 4 shows an example screenshot for Fragment Analyzer and instructions are included in Chapter 4 of each of the instrument user manuals.

      

Figure 4. Capillary Alignment

 Tip 
Learn how to effectively maintain your Agilent parallel capillary electrophoresis systems:
5200, 5300 and 5400 Fragment Analyzer User Manual 
Femto Pulse User Manual 
ZAG DNA Analyzer User Manual 
Oligo Pro II User Manual
Prosize Data Analysis Software User Manual
 
For Research Use only. Not for diagnostic purposes. 
PR # 7000-8228

  • parallel-capillary-electrophoresis
  • how-to-article
  • automated-electrophoresis
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help