Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
  • Home
  • Technical Areas
    Technical Areas
    • AssayMap
    • Atomic Spectroscopy
    • Automated Electrophoresis
    • Bioreagents
    • Cell Analysis
    • Consumables
    • CrossLab Connect
      CrossLab Connect
      • Service Management
      • Smart Alerts
    • Dissolution
    • Gas Chromatography
    • GC/MS
    • LC/MS
    • Liquid Chromatography
    • Molecular Spectroscopy
    • Sample Preparation
    • Software
      Software
      • Chromatography Software
        OpenLab CDS, OpenLab ChemStation, OpenLab EZChrom
      • Data Management Software
        OpenLab ECM, OpenLab ECM XT, OpenLab Server
      • Mass Spectrometry Software
        Data Analysis, Data Acquisition, Productivity App, WalkUp Software
    • Vacuum
  • Applications Areas
    Applications Areas
    • Basic Research
    • Energy and Chemicals
    • Environmental
    • Food and Agriculture
  • Agilent Knowledge Portal
  • Links
    Links
    • Collection of Support Resources
    • Webinar Notifications
    • Contact Support
    • User Contributions
  • Getting Started
    Community Help and Resources
  • Browse
    Browse
    • People
    • Places
    • Your View
    • Your Community Inbox
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Toggle X-Axis Description From Time (Seconds) to Size for 2100 Bioanalyzer System Data
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • -Bioanalyzer
      • DV200 Analysis of FFPE RNA Samples with the 2100 Bioanalyzer System
      • Analysis of Yeast RNA on the Agilent Bioanalyzer
      • Agilent 2100 Bioanalyzer Error: Run Aborted on Port x, Unusual High or Low Voltage
      • RIN Not Calculated or NA for Total RNA Samples in the Agilent Bioanalyzer 2100 Expert Software
      • Wavy Baseline on the Agilent Bioanalyzer
      • Toggle X-Axis Description From Time (Seconds) to Size for 2100 Bioanalyzer System Data
      • Agilent 2100 Bioanalyzer System Chip not Detected
      • Agilent Bioanalyzer Assay Chip Guide
      • How to Install the Agilent 2100 Expert Software and Connect a PC to the Agilent Bioanalyzer
      • Maintenance of the Agilent 2100 Bioanalyzer System Chip Priming Station
      • Can the Agilent 2100 Bioanalyzer be Used to Determine Size of RNA?
      • Connecting Two Agilent Bioanalyzers to One PC
      • Agilent 2100 Bioanalyzer System Run-time Error 50003 upon 2100 Expert Software Launch
      • Region Information not Included in Agilent 2100 Bioanalyzer System PDF Report
      • How to Run the Hardware Diagnostics on the Agilent Bioanalyzer
      • Incorrect Text Placement and Image Size in Agilent Bioanalyzer 2100 Expert Software PDF Reports
    • +Tapestation
    • +Parallel Electrophoresis

Toggle X-Axis Description From Time (Seconds) to Size for 2100 Bioanalyzer System Data

Created by Shweta.S Shweta.S over 1 year ago | Last modified by Geoff Pettitt Geoff Pettitt 8 months ago

This Information Applies To: 2100 Bioanalyzer system, 2100 Expert software, RNA assay, DNA assay, Protein assay 


Issue:

The X-axis description toggle icon in the Agilent 2100 Expert software toolbar (Figure 1) does not change the X-axis from time scale to base pairs, nucleotides, or kilodaltons. Displays X-axis in time scale only.  


Figure 1. 2100 Expert software icon to toggle X-axis description from time to size 


Resolution: 

The common causes of toggle failure include:

  • The software did not identify the peak. To rectify, manually integrate the peak.
  • The software did not automatically assign the upper or lower marker in one or more wells of the 2100 Bioanalyzer assay. As a fail-safe, the software will not present the X-axis in size if a single marker is missing from even one of the samples in the assay. To rectify, manually assign the marker.
  • No marker or sample was loaded onto the well. To rectify, manually save selected data of the ladder and samples that have upper and lower marker.


Manually assign lower or upper marker:

  • Select Peak Table > right-click with the cursor on the target peak > Manually Set Lower (or Upper) Marker (Figure 2).
  • Toggle X-axis to size (Figure 3).

Further details on 'Manual Marker Assignment' can be found starting page 58 of 2100 Bioanalyzer Maintenance and Troubleshooting Guide.

 

Figure 2. Example showing X-axis data defaulted to time.  

 

Figure 3. Example showing X-axis data in size after assigning markers correctly. 

Manually save selected data:

Save the ladder and selected sample data that have the upper and lower marker present to a separate file.

  • Select File > Save Selected Sample (Figure 4) > choose the desired samples > Apply (Figure 5) > Save as new raw data (.xad) file. 
  • Toggle X-axis to size.

 

Figure 4. Choose selected samples. 

     

Figure 5. Save selected samples as a new data file. 

 Tip  For further troubleshooting help, please see Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide. 

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8037

  • Troubleshooting
  • Bioanalyzer
  • Bioanalyzer Software
  • Bioanalyzer Assays
  • 2100 Expert Software
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help