Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles Agilent Parallel Electrophoresis Capillary Array Disposal
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel
  • -Automated Electrophoresis
    • +Bioanalyzer
    • +Tapestation
    • -Parallel Electrophoresis
      • Failed to Build Pressure Error on a Parallel Capillary Electrophoresis Instrument
      • Run Aborted and Target Z Position Error on Parallel Capillary Instruments
      • Analysis of Single Stranded DNA on the Fragment Analyzer
      • How to do a NaOH Flush on an Agilent Fragment Analyzer
      • Agilent Fragment Analyzer Message: Setup Incorrect - Ladder File does not Exist
      • Agilent Fragment Analyzer or Femto Pulse: Run Aborted, Error Moving Syringe Plunger
      • Agilent Parallel Electrophoresis Capillary Array Disposal
      • Re-ordering the Syringe for Cleaning the Reservoir Vent Valve of the Agilent Fragment Analyzer, Femto Pulse, or ZAG
      • DV200 Analysis of FFPE RNA Samples with the Fragment Analyzer
      • Agilent Capillary Array Installation on Parallel Capillary Electrophoresis Instruments
      • Error 97 When Creating PDF Report in ProSize Data Analysis Software
      • How to Clean the Capillary Array Window on Agilent Parallel Capillary Electrophoresis Instruments
      • Reservoir Vent Valve Cleaning for Parallel Capillary Electrophoresis Instruments
      • Full Power Cycle for Agilent Parallel Capillary Electrophoresis Instruments
      • Agilent Parallel Capillary Electrophoresis Compatible Plasticware
      • How to Use Mineral Oil to Prevent Evaporation on Agilent Parallel Capillary Electrophoresis Systems
      • Running Dye Labeled Samples on Agilent Parallel Capillary Electrophoresis Instruments
      • Insert Disk Error in the Agilent Fragment Analyzer Controller Software
      • BAC Method Missing from the Agilent Femto Pulse Software

You are currently reviewing an older revision of this page.

  • History View current version

Agilent Parallel Electrophoresis Capillary Array Disposal

This Information Applies To: Parallel capillary electrophoresis platforms, 5200/5300/5400 Fragment Analyzer, ZAG DNA Analyzer, Femto Pulse, Oligo Pro II 


Issue:

Disposing used capillary arrays 


Resolution: (For How-To Articles)

For capillary array disposal, we recommend performing a deionized water flush on the capillary array before removing it from the instrument. To perform the water flush, open the instruments’ side compartment to replace the gel 2 bottle with a bottle containing water. From the operations tab located on the main screen window, select Add to queue under the Capillary Array-Conditioning menu and perform the water flush for atleast 10 min at 200 psi. Minimum water required to flush 12 -capillary array is ≥ 12 mL, ≥ 25 mL for a 48-capillary array and ≥ 50 mL for a 96-capillary array. However, please note, the water flush might not remove all reagent residues, so the arrays should be treated as hazardous waste when disposing due to traces of intercalating dye and other substances.

 

 Bulb 
Learn how to effectively (maintain -troubleshoot-operate) your (Agilent Product):5200, 5300, and 5400 Fragment Analyzer System Manual, Femto Pulse User Manual, Oligo Pro II User Manual, ZAG DNA Analyzer System Manual 
(link to course learning path, or user manual, include course code) e-learning course on Agilent University/ user manual on Agilent.com/ Technical Note on Agilent.com/ Video on Agilent Community

For Research Use Only. Not for use in diagnostic procedures.

Template Tools - Delete this section after Article Review

Note: Remember to add DE, or RA number and Keywords in the Options tab

 Bulb 
Tips: (Tips, advices, good practices to keep instrument running, same box as Agilen University)

 Bulb 
Learn how to effectively (maintain -troubleshoot-operate) your (Agilent Product):
(link to course learning path, or user manual, include course code) e-learning course on Agilent University/ user manual on Agilent.com/ Technical Note on Agilent.com/ Video on Agilent Community

 Warning 
Caution: (A caution message advises readers that failure to take or avoid a specific action could result in damage to equipment or loss of data.)

 Warning 
Warning: (A warning message advises users that failure to take or avoid a specific action could result in injury or death.)

 Heavy check mark 
Confirmation: (Use only in cases where a specific condition is needed to confirm the resolution)
Privacy Statement
Terms of Use
Contact Us
Site Help