Agilent Community
Agilent Community
  • User
  • Site
  • Search the Community
  • User
Automated Electrophoresis Portal
  • Agilent Knowledge Portal
Automated Electrophoresis Portal
Automated Electrophoresis Articles How to Run the Hardware Diagnostics on the Agilent Bioanalyzer
  • Automated Electrophoresis Articles
  • Automated Electrophoresis Forum
  • More
  • Cancel

Browse the Agilent Knowledge Portal

  • Chromatography
    • Gas Chromatography
    • Liquid Chromatography
  • More
  • Cancel
  • GC/MS
  • More
  • Cancel
  • Spectroscopy
    • Atomic Absorption
    • ICP-OES
  • More
  • Cancel
  • Automated Electrophoresis
  • More
  • Cancel
  • Software and Informatics
    • OpenLab CDS
    • MassHunter
  • More
  • Cancel
  • -Automated Electrophoresis
    • -Bioanalyzer
      • DV200 Analysis of FFPE RNA Samples with the 2100 Bioanalyzer System
      • Analysis of Yeast RNA on the Agilent Bioanalyzer
      • RIN Not Calculated or NA for Total RNA Samples in the Agilent Bioanalyzer 2100 Expert Software
      • Toggle X-Axis Description From Time (Seconds) to Size for 2100 Bioanalyzer System Data
      • Agilent 2100 Bioanalyzer System Chip not Detected
      • Agilent Bioanalyzer Assay Chip Guide
      • How to Install the Agilent 2100 Expert Software and Connect a PC to the Agilent Bioanalyzer
      • Maintenance of the Agilent 2100 Bioanalyzer System Chip Priming Station
      • Can the Agilent 2100 Bioanalyzer be Used to Determine Size of RNA?
      • Connecting Two Agilent Bioanalyzers to One PC
      • Agilent 2100 Bioanalyzer System Run-time Error 50003 upon 2100 Expert Software Launch
      • Region Information not Included in Agilent 2100 Bioanalyzer System PDF Report
      • How to Run the Hardware Diagnostics on the Agilent Bioanalyzer
      • Incorrect Text Placement and Image Size in Agilent Bioanalyzer 2100 Expert Software PDF Reports
    • +Tapestation
    • +Parallel Electrophoresis

How to Run the Hardware Diagnostics on the Agilent Bioanalyzer

Created by Geoff Pettitt Geoff Pettitt 11 months ago | Last modified by Shweta.S Shweta.S 1 month ago

This Information Applies To: 2100 Bioanalyzer, 2100 Expert software


Issue:

Running the hardware diagnostics on the Agilent 2100 Bioanalyzer system can help to investigate whether the instrument performs as expected.


Steps to follow:

Required material

  1. An empty clean unused chip of any type (DNA, RNA, or protein)
  2. The test chip kit (p/n G2938-68300) consisting of an electrode / diode test chip and an autofocus test chip
 Tip  Tip: Verify that the test chip kit is within the expiry date and the test chips have been stored in their original bag. Avoid touching the underside of the test chips.

The Bioanalyzer Troubleshooting and Maintenance Guide provides an overview of the diagnostics in the Hardware Diagnostics section.

  1. Start the Agilent 2100 Expert software.
  2. Access the hardware diagnostic tests by selecting the Diagnostics tab in the Instrument context of the 2100 Expert software (Figure 1).
  3. In case multiple 2100 Bioanalyzer instruments are connected to the PC, highlight the appropriate instrument in the tree view. 
  4. Select any of the hardware tests from the list or choose Select All to run all tests.
  5. Select Start and follow the instructions as given by the 2100 Expert software. 
  6. If there are any failures, repeat the failed tests in isolation.
  7. If the failures persist, send the diagnostic test results file to Agilent technical support in the country where the instrument is located. The results of diagnostic tests are stored as .xdy files in C:\Program Files (x86)\Agilent\2100 bioanalyzer\2100 expert\diagnosis. Alternatively, create a Support Package by clicking Help > Create Support Package.... The File Collect dialog box will appear. Select all check boxes to avoid useful information being missed. Clicking Collect will generate a support package and save it to your desktop (Figure 2). This support package will contain the diagnostic results and should be sent to Agilent technical support via email. 

 

Figure 1. Bioanalyzer Diagnostics tab

Figure 2. How to create a support package in the 2100 Expert software

Resolving common test failures

Short circuit test

  • If the failure occurs after electrode pin set cleaning, make sure that the pin set is completely dried before rerunning the test. Remove the pin set from the electrode cartridge and dry both with oil-free compressed air thoroughly. Alternatively, let the pin set dry overnight.
  • If the pin set wasn't cleaned before the diagnostics, perform the maintenance of the electrode cartridge as described in the Bioanalyzer Troubleshooting and Maintenance Guide on page 155.
  • Ensure that the humidity level in the environment is below 50% and the ambient temperature is 25 °C.

Optics test

Clean the focusing lens with lens-grade, lint-free tissue paper, and isopropanol before rerunning the test.

  Tip
Learn how to effectively maintain your Bioanalyzer:
Agilent 2100 Bioanalyzer System Maintenance and Troubleshooting Guide

For Research Use Only. Not for use in diagnostic procedures.

PR7000-8237

  • Bioanalyzer Instruments
  • Troubleshooting
  • Bioanalyzer
  • Bioanalyzer Software
  • 2100 Bioanalyzer
  • 2100 Expert Software
  • automated-electrophoresis
  • Share
  • History
  • More
  • Cancel
Was this helpful?
  • Yes
  • No
Thank you for your feedback!

Agilent Community Feedback

Agilent Community Feedback

×
We are glad this was helpful! We are sorry this was not helpful. If you still need assistance please create a community post or contact support. To help us improve, please provide any additional feedback. For full details of how we will treat your information, please view our privacy policy.
Submit Cancel
Submit Cancel

Contact Agilent Support

 Still Need Assistance?

 Contact Agilent Support >

Related
Recommended for you
Privacy Statement
Terms of Use
Contact Us
Site Help